Recently Lasersoft has been hawking around these "Silverfast Advanced Targets (Part 2)", which they claim:
"With our newly designed Part 2 Targets, users receive three times more measuring fields, while even more precise measurements provide a significantly larger color spectrum. The color profile is further supported by additional dark, grey and pastel toned color fields, which are supported by significantly truer measured values."
Do any colour gurus out there (a) have any idea what Lasersoft are talking about, and (b), if so, is there any tangible benefit likely over standard IT8 calibration ?
Hawking? Not really, it's not their invention; it's based on a new spec (ISO 12641-2). But they do offer it. And yes, I've tested it.
Very initial impressions of two targets used to build a reflective scanner profile (Epson V750).
One made with the IT8 original from LaserSof: IT8.7/2-1993 2015:03
One made with the new target: ISO:12641-2 2017:11
The newer target has a larger color gamut and gamut volume which is to be expected. But this expectation is good! The gamut of a profile cannot be any larger than the target used to build it and thus, this new target is a major improvement there compared to the original as seen below. Now what I'm still testing (or waiting to hear back from LaserSoft) is why using the two profiles produces so little actual difference in two scans of my Gamut Test File printed on a 3880. I'd have expected the wider gamut input profile would produce a wider gamut scan and I'm not sure why the deltaE differences of the two are so small.
--------------------------------
dE Report
Number of Samples: 208500
Delta-E Formula dE2000
Overall - (208500 colors)
--------------------------------------------------
Average dE: 0.57
Max dE: 6.01
Min dE: 0.00
StdDev dE: 0.51
Best 90% - (187649 colors)
--------------------------------------------------
Average dE: 0.44
Max dE: 1.21
Min dE: 0.00
StdDev dE: 0.30
Worst 10% - (20851 colors)
--------------------------------------------------
Average dE: 1.73
Max dE: 6.01
Min dE: 1.21
StdDev dE: 0.52
--------------------------------------------------
--------------------------------------------------
What you see is the deltaE differences in the two are tiny and largely invisible. I'd expect that colors that fall outside the gamut of the 'standard' scanner profile would show larger dE values than using the Advanced profile. Note I'm using Standard and Advanced to name the two targets and thus the profiles they create.