Joe: What you are seeing with the i1Pro is indeed the limitations of its measurement capabilities. From our results, the i1Pro has a real-world noise floor of between 0.3 - 0.4 cd/m2. At lower luminance values, i1Pro numbers are more random than real. When you raise the blackest black from
0.11 to 0.35, you are getting more signal, less noise, and the green casts to the shadows go away. (In all likelihood, the true monitor black point is 0.05 - 0.1 cd/m2 higher than what the i1Pro is reporting).
CEDP increases measurement integration time in darker tones above what other (non-Argyll) programs do. The price you pay for BasICColor Display's speed is inaccurate shadow readings. Measure the same monitor with a DTP-94, and BasICColor should give more consistent black levels. The white point will be all over the place, however.
I spent more time with i1Profiler over the weekend. When choosing "native" contrast ratio, it gives a higher black point than does CEDP's relative black, no matter whether an i1Pro or i1D2 is used. At the black level CEDP chooses, a good monitor (e.g. Eizo CG243W) is capable of differentiating RGB (1, 1, 1) from (0, 0, 0). By setting a higher level, i1Profiler both gets around the limitations of the i1Pro and allows faster measurement times. Whether this works for you depends on how black you like it.
The absence or presence of color casts at the lowest luminance levels appears to be a function of the sensor being used. One calibration suite (
CalPC) we are evaluating supported our PR-730 spectroradiometer. With a monitor that it could talk to via DDC, it produced dead-neutral shadows down to 0.15 cd/m2.
BasICColor Display also supports fancy equipment, just in case anyone has a spare KM CS-2000 floating around. Once we get a Discus in-house, we will check BasICColor's performance down in the mud.