Raw & Post Processing, Printing => Colour Management => Topic started by: Christopher on July 08, 2011, 04:03:10 am

Title: Eizo CG 301 / display validation?
Post by: Christopher on July 08, 2011, 04:03:10 am
Hi,

I just wanted to ask what Delta-E values are still good for a display like the Eizo CG301. (CIE2000 , or 1994 or 1976 all welcome)

Just want to make sure the thing is working as it should. Thanks for the help.

Chris
Title: Re: Eizo CG 301 / display validation?
Post by: tho_mas on July 08, 2011, 05:00:13 am
Hi Christopher.

Keep in mind profile validation only reflects the Delta E of the current monitor state in relation to the values stored in the monitor profile - both based on the values measured with the same colorimeter (or spectrometer).
Still validation is useful to check particularly the greyscale (in addition to a visual inspection).
My CG241W has been running about 11.500 hours and the greyscale is in the range of 0.1-0.4 Delta E.
Colors are in the range of 0.1 - 0.8 Delta E ... only the dark blue patch (RGB 0/0/127) shows a max. Delta E of 1.7.

I wouldn't worry too much about these values... the visual match to actual prints is much more important. But that requires a lot more than just a good monitor with good calibration.
Title: Re: Eizo CG 301 / display validation?
Post by: Czornyj on July 08, 2011, 05:15:06 am
Exactly!
Higher dE values might just reflect the behavior of the sensor. Here's an example of i1pro, that generally has problems with noise in dark patches readings:
(http://members.chello.pl/m.kaluza/dscs4.jpg)

Another example shows the influence of heat - same spectrophotometer, but I left it on display for a couple of minutes and repeated validation:
(http://members.chello.pl/m.kaluza/dscs5.jpg)

And for comparison - basICColor DISCUS example results:
(http://members.chello.pl/m.kaluza/dscs2.jpg)