Did you measure the calibration strips manually too as an extra check? Do you have some old ones lying around to see whether the heads lay down less ink or the printer's densitometer/calibration software is acting weird?
Ernst, did the manual measurements of old vs new calibrations. The new calibration shows less density for all colors except for black and yellow, which corresponds to the overall decrease in gamut based on ColorThink's plots. They are about 0.5 L* points higher. A and B values also smaller in the newer target so less saturation. Side by side I can just make out the differences.
I'm not sure if this could be considered a densitometer issue. The differences are within 2dE2k, so its within Canon's specifications.
Lump it! The densitometer should have a repeatability of less than 0.5dE. Anything more than 1dE means re-profiling should be done.
a head that is laying down less ink should be more prone to banding issues too. If nozzles fail they are replaced on the Canon heads checked by another sensor than the densitometer I guess but as written it is more likely that banding shows if both nozzles and the sensor fail.
The basic voltage to the heads going down but still within the specs to keep error reports out? The gamut decreasing evenly over the boundaries or on specific hues? The old calibration target prints may tell that too.
That's a brilliant observation. I was indeed struggling with slight banding issues. After multiple head adjustments (automatic, initial head adjustment as though installing a new head, and manual too) I see most of the issue goes away, but there is slight residual banding in warm browns, cyans, blues, purples. There might be some in the oranges and reds but I can't see it. Very very slight.
I showed the banding - its clearer on a printer profiling target - to the Canon tech who came the other day. He took quite a while to see it.
I don't get any error reports. What do you recommend to do? Should I ask the Canon tech to check the power supply? If that's the issue changing the heads alone won't solve the problem? I'll remember to ask him to check the head management and multi sensor anyway.