I recently purchased an i1Pro Xtreme bundle, upgraded it to i1Profiler Publish software, and before exploring its 2nd iteration profiling, I decided to investigate the spectrophotometer repeatability varying targets between scrambled and non-scrambled patches and between 7.3 and 10mm width printed patches. For the first moment, I printed four 400-patch targets on normal paper, waited overnight to dry and scanned each of them twice, with a moderate to fast movement speed (around 2 seconds per row) through the scanning ruler.
Comparing each measurement to the other of the same target, i1Profiler Data Analysis gave me the following results:
-- 10mm patch width and scrambled patches:
Average DeltaE (deltaE2000) results [Measured (Patches)]
Total: 0.10
Lowest 90%: 0.09 (360)
Highest 10%: 0.22 (40)
Standard deviation results
Total: 0.06
Lowest 90%: 0.04 (360)
Highest 10%: 0.04 (40)
Maximum results
Max DeltaE all patches: 0.33
Max DeltaE in the lowest 90%: 0.18 (360)
-- 7.3mm patch width and scrambled patches:
Average DeltaE results
Total: 0.26
Lowest 90%: 0.20 (360)
Highest 10%: 0.79 (40)
Standard deviation results
Total: 0.25
Lowest 90%: 0.11 (360)
Highest 10%: 0.44 (40)
Maximum results
Max DeltaE all patches: 2.19
Max DeltaE in the lowest 90%: 0.49 (360)
-- 10mm patch width and non-scrambled patches:
Average DeltaE results
Total: 0.11
Lowest 90%: 0.10 (360)
Highest 10%: 0.27 (40)
Standard deviation results
Total: 0.07
Lowest 90%: 0.05 (360)
Highest 10%: 0.06 (40)
Maximum results
Max DeltaE all patches: 0.47
Max DeltaE in the lowest 90%: 0.21 (360)
-- 7.3mm patch width and non-scrambled patches:
Average DeltaE results
Total: 0.14
Lowest 90%: 0.11 (360)
Highest 10%: 0.45 (40)
Standard deviation results
Total: 0.13
Lowest 90%: 0.07 (360)
Highest 10%: 0.12 (40)
Maximum results
Max DeltaE all patches: 0.86
Max DeltaE in the lowest 90%: 0.32 (360)
--
Since the best statistics were reached with wider patches for the scrambled target and non-scrambled target for the narrower patches, I had made once again the first and the forth measurements with a slower scanning speed through the ruler (around 4 seconds per row), and got the following results:
-- 10mm patch width and scrambled patches:
Average DeltaE results
Total: 0.12
Lowest 90%: 0.11 (360)
Highest 10%: 0.25 (40)
Standard deviation results
Total: 0.06
Lowest 90%: 0.05 (360)
Highest 10%: 0.04 (40)
Maximum results
Max DeltaE all patches: 0.40
Max DeltaE in the lowest 90%: 0.21 (360)
-- 7.3mm patch width and non-scrambled patches:
Average DeltaE (DeltaE2000) results
Total: 0.12
Lowest 90%: 0.10 (360)
Highest 10%: 0.32 (40)
Standard deviation results
Total: 0.09
Lowest 90%: 0.06 (360)
Highest 10%: 0.07 (40)
Maximum results
Max DeltaE all patches: 0.53
Max DeltaE in the lowest 90%: 0.25 (360)
As someone could see, these two last measurements gave me similar numbers. As it became evident scanning speed really matters, I do not mind scanning them slower for the sake of more precision and better repeatability. But it would be great if I could maintain the 7.3mm width patches as they produce much less printed targets (for huge number of patches) and less scanning work too. But again, I do not want to loose any necessary precision of repeatability, but do not mind loosing unnecessary repeatability. Does anyone could tell me if the last measurement (7.3mm non-scrambled) has sufficient good repeatability for 2000-4000 patch targets? In other words, if their average deltaE(2000) are smaller than the small patch-to-patch differences in those thousand-numbered targets?