Raw & Post Processing, Printing => Colour Management => Topic started by: shayaweiss on June 21, 2010, 01:58:36 pm

Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 21, 2010, 01:58:36 pm
Does anybody know whether SpectraView II uses software matrices to correct for wide gamut monitors when used in connection with regular colorimeters like the i1d2 and dtp94?

thanks in advance

shaya
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 21, 2010, 02:55:49 pm
I just got confirmation via email from NEC Display Service & Support that SpectraView II does NOT make any matrices correction for wide gamut. The sensor has to support the wide gamut.

There are only three (or four) sensors that supports wide gamut,
 - the NEC custom mated i1d2 and
 - the Spyder 3.
 - Eizo's and Quato's bundle of matched screen & colorimeter using the dtp94
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 21, 2010, 03:37:17 pm
Quote from: shayaweiss
I just got confirmation via email from NEC Display Service & Support that SpectraView II does NOT make any matrices correction for wide gamut. The sensor has to support the wide gamut.

There are only three (or four) sensors that supports wide gamut,
 - the NEC custom mated i1d2 and
 - the Spyder 3.
 - Eizo's and Quato's bundle of matched screen & colorimeter using the dtp94
NEC specifically told me the i1d2 supports wide gamut and its listed in their manual as doing so..

Not only that.. it actually works and profiles the wide gamut as you can clearly see from the attachments..

After you buy your NEC puck.. it will be interesting to see you run it side by side to your current i1d2 puck and show us the difference..

Still curious about the differences of the Euro vs. America vs. Asian spec monitors..

[attachment=22722:i1d2.JPG]  [attachment=22723:i1d2a.JPG]
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 21, 2010, 04:50:06 pm
Quote from: Steve Weldon
NEC specifically told me the i1d2 supports wide gamut and its listed in their manual as doing so..

I think it just means they work, but those four work better. It again boils down to your needs. I would like to hear from the specialists, what the actual visual difference should be between a calibration with or without a puck corrected for wide gamut. All I can tell you is, that as long as you have only one monitor, and your eyes and/or print to screen match are not that demanding, or you are printing out of house, than you might not notice. I can't tell you all the specific situation that exist of way you might not notice. I did not see it, simply because I never saw something else, or better. But everything changed with two monitors, my eyes couldn't get used. As they had not the same tint, and not to the same degree, I suddenly realized that maybe the tint could go away completely. The only remedy was, you remember,  a colorimeter corrected for wide gamut in the software or the hardware. I am sure I could have used a spectrometer, with similar results, though some say that the colorimeters do better with displays.

Quote
Not only that.. it actually works and profiles the wide gamut as you can clearly see from the attachments..

Yes, it works, also with my monitors, all the curves and diagrams, but somehow it's different, uncorrected, unfiltered for wide gamut. Also you could run a validation, but for that you have to use a different instrument and/ or software to make sens. From experience, I can tell you, that there are variances not only between instruments, but also between different software.

Quote
After you buy your NEC puck.. it will be interesting to see you run it side by side to your current i1d2 puck and show us the difference..

I am already working on that, but it is very difficult to get one at a reasonable price and have it shipped overseas. Suggestions anyone?

Quote
Still curious about the differences of the Euro vs. America vs. Asian spec monitors..
I don't know about Asia. The MultiSync from Europe and America are identical, as stated in these pages. If I remember correctly what was said (for example here http://luminous-landscape.com/forum/index....showtopic=42040 (http://luminous-landscape.com/forum/index.php?showtopic=42040) and here http://luminous-landscape.com/forum/index....howtopic=44312) (http://luminous-landscape.com/forum/index.php?showtopic=44312)), the difference is only in the Spectra View models. In America they are the same as the MultiSync but ship with the SpectraView software and the NEC i1d2 puck corrected for wide gamut. In Europe the SpectraView are handpicked among the best monitors, they come with a hood and a version for NEC of BasICColor display 4, and therefore they are also much more expensive.
Title: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on June 21, 2010, 05:49:06 pm
Quote from: shayaweiss
There are only three (or four) sensors that supports wide gamut,
 - the NEC custom mated i1d2 and
 - the Spyder 3.
 - Eizo's and Quato's bundle of matched screen & colorimeter using the dtp94

Spyder3? I wouldn't count on it  

Eizo & Quato use correction matrix for in ColorNavigator & iColor display software - it works in combination with trusty DTP94, that has decent inter instrumental agreement.

NEC uses custom mated colorimeters, and that's why they don't use correction matrix in the software.

Quote from: shayaweiss
I don't know about Asia. The MultiSync from Europe and America are identical, as stated in these pages. If I remember correctly what was said (for example here http://luminous-landscape.com/forum/index....showtopic=42040 (http://luminous-landscape.com/forum/index.php?showtopic=42040) and here http://luminous-landscape.com/forum/index....howtopic=44312) (http://luminous-landscape.com/forum/index.php?showtopic=44312)), the difference is only in the Spectra View models. In America they are the same as the MultiSync but ship with the SpectraView software and the NEC i1d2 puck corrected for wide gamut. In Europe the SpectraView are handpicked among the best monitors, they come with a hood and a version for NEC of BasICColor display 4, and therefore they are also much more expensive.

All NEC panels are physically the same - the only difference is, that hand picked european Spectraview panels have firmware modification, and NEC EU blessing - so they can be hardware calibrated with Spectraview display a.k.a. basICColor. In the past there was a method to turn x90 series panel into "Spectraview mode", using simple key combination - now it's more complicated.

US Spectraview II represents more democratic approach, and calibrates all x90, PA and P series NEC displays.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 21, 2010, 05:55:27 pm
Quote from: Czornyj
Spyder3? I wouldn't count on it  

Eizo & Quato use correction matrix for in ColorNavigator & iColor display software - it works in combination with trusty DTP94, that has decent inter instrumental agreement.

NEC uses custom mated colorimeters, and that's why they don't use correction matrix in the software.

Exactly.

And as you know, I recently found out through trial and error.

Only, I have no experience with the spyder3. I didn't even hear of it a lot. I mentioned it because the NEC display technician told me in his e-mail, that there where two instruments that are corrected for wide gamut monitors that he knew of, their own i1d2 and the spyder3 .  But why is it not dependable? I was thinking of it, as it is much easier to get than the NEC puck. But if you say so, I will most certainly stay away from it. Just maybe could you give us some hints or details about whats not so good about it.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 21, 2010, 05:58:37 pm
Here is the above mentioned email (with permission)

-----Original Message-----
From: NEC Display Service & Support [mailto:techsupport@necdisplay.com]
Sent: Monday, June 21, 2010 9:39 PM
To: #######################
Subject: RE: NEC sensor matrix VS SpectraView II matrix
(#8477-180783712-8459)

Shaya,

Thank you for contacting NEC Display
The calibration software does not make any matrices correction for wide gamut The sensor has to support the wide gamut.

There are only two sensors that I'm aware of that supports wide gamut, Ours and the Spyder 3

You can read more about Spectraview and download the users manual under the downloads tab at:
http://www.necdisplay.com/supportcenter/mo...ectraview2/faq/ (http://www.necdisplay.com/supportcenter/monitors/spectraview2/faq/)

M. Daly
NEC Display Service & Support
Telephone: 1-800-632-4662
Email: techsupport@necdisplay.com
Title: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on June 21, 2010, 06:35:03 pm
Quote from: shayaweiss
Exactly.

And as you know, I recently found out through trial and error.

Only, I have no experience with the spyder3. I didn't even hear of it a lot. I mentioned it because the NEC display technician told me in his e-mail, that there where two instruments that are corrected for wide gamut monitors that he knew of, their own i1d2 and the spyder3 .  But why is it not dependable? I was thinking of it, as it is much easier to get than the NEC puck. But if you say so, I will most certainly stay away from it. Just maybe could you give us some hints or details about whats not so good about it.

http://www.lumita.com/site_media/work/whit...xrite-wp-3a.pdf (http://www.lumita.com/site_media/work/whitepapers/files/xrite-wp-3a.pdf)

In theory, Spyder3 with 7 color filters should match CIE XYZ function better. In practice - it seems there's still something wrong.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 21, 2010, 06:59:08 pm
Quote from: Czornyj
http://www.lumita.com/site_media/work/whit...xrite-wp-3a.pdf (http://www.lumita.com/site_media/work/whitepapers/files/xrite-wp-3a.pdf)

In theory, Spyder3 with 7 color filters should match CIE XYZ function better. In practice - it seems there's still something wrong.

Thank you for this highly informative document by Karl Lang! I understand that a custom filtered colorimeter will be the best choice. But I didn't read any hint to the 7 filters, or why it should not work.
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 22, 2010, 01:38:03 am
Quote from: shayaweiss
Yes, it works, also with my monitors, all the curves and diagrams, but somehow it's different, uncorrected, unfiltered for wide gamut. Also you could run a validation, but for that you have to use a different instrument and/ or software to make sens. From experience, I can tell you, that there are variances not only between instruments, but also between different software.
I'm not buying my profiles are uncorrected.  They perfectly match the physical color checker on every shade.. and I mean perfectly.

I can accept you've had problems with your Eizo.. but you haven't yet used your i1d2 with the NEC.   And when I wrote NEC they told me (and listed in their instruction manuals and on-line text) the i1d2 was compatible with their wide gamut monitors.  

Curious though.. I know people using the Munki, most are using the i1d2, some the SVII.. are they creating different profiles?  Probably.  Probably small differences.  The question is:  Is one profile more correct than the other.  How would you test it?  I test against a physical color checker.  Anyone else?
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 22, 2010, 01:40:20 am
Quote from: Czornyj
All NEC panels are physically the same - the only difference is, that hand picked european Spectraview panels have firmware modification, and NEC EU blessing - so they can be hardware calibrated with Spectraview display a.k.a. basICColor. In the past there was a method to turn x90 series panel into "Spectraview mode", using simple key combination - now it's more complicated.

US Spectraview II represents more democratic approach, and calibrates all x90, PA and P series NEC displays.
Thanks.  This helps clear things up.

What do Europeans use for calibration if they don't have a reference standard?
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 22, 2010, 05:50:28 am
Quote from: Steve Weldon
And when I wrote NEC they told me (and listed in their instruction manuals and on-line text) the i1d2 was compatible with their wide gamut monitors.

Steve,

I don't know what they told you, but it is not what Karl Lang writes (see the document above), it is not what I got as information from NEC (see email above), it is also not what Czornyj and others have wrote here in this forum (see quotes above), and it is not what you can read on the Argyll/dispcalGUI information (see quote http://luminous-landscape.com/forum/index....t&p=372083) (http://luminous-landscape.com/forum/index.php?s=&showtopic=44078&view=findpost&p=372083)), and it does also not meet what I have read from other specialists across the Web.

I'm not a specialist, and you know that, but I try to understand. So what can NEC mean, when they nonetheless state in their instruction manual that the i1d2 and other unfiltered colorimeters are compatible? (BTW is there any instrument that is not in that list?) I think that's what I wrote. They mean to say that it works, just not so well.

Also, as I said, I can visually tell the difference between a calibration and profile made with a corrected colorimeter or an uncorrected one. You have to try it, to see it. Alternatively, you could try a software that corrects for wide gamut, like Quato's iColor Display 3, though the latter would not be able to do hardware calibration. As mentioned earlier, the correction can be made either in the hardware or the software.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 22, 2010, 06:01:06 am
Quote from: Steve Weldon
I test against a physical color checker.

What is your procedure?
Title: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on June 22, 2010, 06:28:25 am
Quote from: Steve Weldon
What do Europeans use for calibration if they don't have a reference standard?

European Spectraview display profiler has correction matrix for NEC display spectra and popular colorimeters, so it's the same approach as Eizo ColorNavigator, and Quato iColor display.
NEC Display America choosed to build custom mated colorimeter rather than make software matrix correction. The Spectraview II checks the sensor type, and when it's not NEC colorimeter or spectrophotometer it creates profiles using factory avaraged measurements taken during production. See Spectraview II Edit>Preferences>ICC Profile tag.

Testing the profile quality against CC24 gives only overall idea of how the profile works, and as a matter of fact it's impossible to get 100% visual connection - the spectra of the displays backlight and the spectra of our lighting is different, and so on...
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 22, 2010, 11:05:22 am
Quote from: shayaweiss
Steve,

I don't know what they told you, but it is not what Karl Lang writes (see the document above), it is not what I got as information from NEC (see email above), it is also not what Czornyj and others have wrote here in this forum (see quotes above), and it is not what you can read on the Argyll/dispcalGUI information (see quote http://luminous-landscape.com/forum/index....t&p=372083) (http://luminous-landscape.com/forum/index.php?s=&showtopic=44078&view=findpost&p=372083)), and it does also not meet what I have read from other specialists across the Web.

I'm not a specialist, and you know that, but I try to understand. So what can NEC mean, when they nonetheless state in their instruction manual that the i1d2 and other unfiltered colorimeters are compatible? (BTW is there any instrument that is not in that list?) I think that's what I wrote. They mean to say that it works, just not so well.

Also, as I said, I can visually tell the difference between a calibration and profile made with a corrected colorimeter or an uncorrected one. You have to try it, to see it. Alternatively, you could try a software that corrects for wide gamut, like Quato's iColor Display 3, though the latter would not be able to do hardware calibration. As mentioned earlier, the correction can be made either in the hardware or the software.
I'm not sure what qualified as a "Specialist.."  I'm just a guy who has been using color profiled monitors for well over a decade and have helped others set up dozens of systems.  I have 'some' experience, but nothing that qualifies me as a "specialist" and I'm naturally skeptical of anyone making that claim.

I'm just not buying NEC puts a colorimeter in their catalog and instruction manuals as "Compatible, but just not so well.."  That's silly.  I do believe it's possible all colorimeters listed meet minimum tolerances and some 'might' hold tighter (but not different) tolerances.. but now we're taking the word of "specialists" on the net.. already explained how I feel about that.

You have the NEC puck as well as the 1id2?  And you've used them on a known good NEC monitor?   I'm sorry, but from what you've listed of your experiences you can't really be sure of anything.  You've had a couple questionable Eizo monitors, many different programs, and a testing strategy that (to me) doesn't appear logical in the sense that you were trying to pin down specific things that work or not.. only if it worked.  There's a difference.

And no.. anyone running an NEC monitor and not using SVII is defeating the main advantages of owning an NEC monitor with DDC support.. I hope you don't plan on using anything but SVII.. if you do good luck.

Lets look at it this way.  I've been using color profiled monitors for many years.. on matched monitors and single workstations.  I know what the colors should look like and if they weren't correct I'd know and my entire workflow wouldn't work.  sRGB emulation is new to me.. but my colors are holding across all gamuts providing I work my entire workflow with the same gamut/profile.  Test images from Phase One and others (images worked up to show the color checker and more just for this purpose) and a physical color checker are perfect.  IF I did run another puck and got other colors.. they wouldn't be right.  Or the differences would be so small only the colorimeter would see it.. and not me.

Do I believe you're having issues?  Yes.  But I'm guessing a lot of them are a.  Self induced  or  b.  You're worrying about things you shouldn't yet be worried about concerning the NEC monitor.

Now.. when you get your NEC monitor and you have both the NEC puck and the i1d2 puck.. and you run them side by side and tell me you're getting 'significant' visible differences.. then we'll talk more.  Until then you can keep reading all the "specialist" talk on the net and wringing your hands and anticipating issues you might or might not have when you get your new monitor.

How do you know a color profile is correct?  You compare it to references.  A color checker print out in a light tent compared to the real thing gets you really close.

There are two reasons you color profile.

1.  So the prints you make, either from your own printer or a print shop, look as they should.

2.  So your files are the same as a standard that you can share across the profession.

If they do 1&2 above.. to the tolerances needed for your work.. then you're color profiled.  

It's going to be very interesting when you get your NEC monitor..
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 22, 2010, 11:11:23 am
Quote from: Czornyj
European Spectraview display profiler has correction matrix for NEC display spectra and popular colorimeters, so it's the same approach as Eizo ColorNavigator, and Quato iColor display.
NEC Display America choosed to build custom mated colorimeter rather than make software matrix correction. The Spectraview II checks the sensor type, and when it's not NEC colorimeter or spectrophotometer it creates profiles using factory avaraged measurements taken during production. See Spectraview II Edit>Preferences>ICC Profile tag.

Testing the profile quality against CC24 gives only overall idea of how the profile works, and as a matter of fact it's impossible to get 100% visual connection - the spectra of the displays backlight and the spectra of our lighting is different, and so on...
Great explanation.. This explains why a 1id2 'could' be very accurate, or slightly less accurate but close, with a NEC monitor using SVII.. and not work well at all on another brand..

It also makes me think an actual colorimeter that can measure without the need of a corrected profile would be the 'ideal' option.. and probably the most accurate consistently.  But also a corrected puck could still be very accurate assuming the person has an "average" sample..

And yes.. 100%.. never gonna happen..   Totally agree.   There comes a point where your experience and workflow matter more than numbers..
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 22, 2010, 11:30:51 am
I don't need decades of experience to see that a monitor has color cast or doesn't much a second monitor. I do also not need that kind of background, to see that a change of Instrument and/or software does change the outcome, and that's no anticipatory talk.

For the rest, as I told you, your critics are pointed at the wrong person, I am not a specialist and have no experience. You say different then the others (I gave you a whole list already), I don't mind, you still might be right, but chances are slim. That is what my professional experience is able to tell, as I spent decades in humanities. For the rest I am just not the right interlocutor.

I would be very pleased to follow your exchange with them. As my reason for writing here is to understand and learn. If you want to knock me down with years of experience, even though what you say goes against this or that view expressed here and there by others with decades of experience, than there is no reason to talk anymore.

Again, I would be pleased if these technical issues could be addressed in a civilized manner from now on.

thank you.

P.S.: Very interesting that you do not address your points with Czornyj, whom I was quoting among others.
Title: SpectraView II matrices for wide gamut monitors?
Post by: AndreaPiaggesi on June 22, 2010, 11:38:12 am
Quote from: Czornyj
NEC Display America choosed to build custom mated colorimeter rather than make software matrix correction. The Spectraview II checks the sensor type, and when it's not NEC colorimeter or spectrophotometer it creates profiles using factory avaraged measurements taken during production. See Spectraview II Edit>Preferences>ICC Profile tag.

It's interesting, so the Spectraview software should use the matrix corrections with the i1Display2 colorimeter.
Sorry for the question buy why the Spyder3 is not recommended?


I live in Europe and I can't afford the original NEC colorimeter (the cost of shipping and tax is too high, it's better to get the Spectraview reference version in this case) so I've two ways:
1) the Spectraview II software with the Spyder3 colorimeter;
2) the Spectraview II software with the i1Display2 colorimeter and correction matrix.

I already own the i1Display2 and I'm thinking to sell it for the Spyder3.

Thanks!!
Title: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on June 22, 2010, 12:05:14 pm
Quote from: AndreaPiaggesi
It's interesting, so the Spectraview software should use the matrix corrections with the i1Display2 colorimeter.
Sorry for the question buy why the Spyder3 is not recommended?


I live in Europe and I can't afford the original NEC colorimeter (the cost of shipping and tax is too high, it's better to get the Spectraview reference version in this case) so I've two ways:
1) the Spectraview II software with the Spyder3 colorimeter;
2) the Spectraview II software with the i1Display2 colorimeter and correction matrix.

I already own the i1Display2 and I'm thinking to sell it for the Spyder3.

Thanks!!

Spyder3 seem to have poor inter instrumental agreement, and I simply wouldn't trust that sensor. It has been discussed in this topic, post #17, 18:
http://luminous-landscape.com/forum/index....showtopic=39210 (http://luminous-landscape.com/forum/index.php?showtopic=39210)
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 22, 2010, 01:23:41 pm
The off the shelf EyeOne-Display will work. As will other Colorimeters not filtered for this kind of unit. The target white point you ask for, and what you get will likely be off. In tests I did, the differences amounted to about CCT 500K. Its not really a huge big deal because the values you ask for are not as important as getting a White Point that produces a visual match to the print next to the display. Yes, the mated colorimeter is preferable! But its not a game changer either.
Title: SpectraView II matrices for wide gamut monitors?
Post by: annamaerz on June 22, 2010, 03:10:56 pm
Quote from: Czornyj
Spyder3 seem to have poor inter instrumental agreement, and I simply wouldn't trust that sensor. It has been discussed in this topic, post #17, 18:
http://luminous-landscape.com/forum/index....showtopic=39210 (http://luminous-landscape.com/forum/index.php?showtopic=39210)

from the texte you refere to, Czorny, I understand that the regular i1d2 is even a tiny little bit worse off, than the spyder3.

Quote
Inter instrument agreement between at least 20 units:

- EyeOne Display 2 - max 18E, mean 8 E
- Spyder 3 Elite - max 15 E, mean 7 E
- DTP94b - max 3 E, mean 1,5E
- EyeOne Pro Rev D - max 3 E, mean 2 E

In the summery there is this recommandation:
Quote
All Colorimeters need an additional correction matrix on wide gamut and white LED displays. Additionally, 2 of the 3 evaluated colorimeters suffer from poor inter instrument agreement. The DTP94 (not available from Xrite any more) performed close to the EyeOne Pro in regards of inter instrument agreement and with a correction for wide gamut, it showed the best results even on the displays with lower black luminances.

Is there such a thin as a DTP94b that has been hardware corrected? I don't think so, right? Or is there any additional trick, software or whatever that can be done to correct it, when using SpectraViewII? I guesse here also, the answer is no.
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 22, 2010, 03:13:07 pm
As far as I know, the DTP94 was not optimized for wide gamut displays.
Title: SpectraView II matrices for wide gamut monitors?
Post by: annamaerz on June 22, 2010, 03:17:02 pm
Quote from: digitaldog
The off the shelf EyeOne-Display will work. As will other Colorimeters not filtered for this kind of unit. The target white point you ask for, and what you get will likely be off. In tests I did, the differences amounted to about CCT 500K. Its not really a huge big deal because the values you ask for are not as important as getting a White Point that produces a visual match to the print next to the display. Yes, the mated colorimeter is preferable! But its not a game changer either.

 After calibration and before profiling, I always tweak the white point to match my paper white (paper white against in viewing booth, against white on the screen). I take a new reading of that white point and set is as a new target white point.  What I noticed is that I rarely correct for more than 40K ! But these 40K difference along with some magenta-green cast correction does make a big visual difference.
Title: SpectraView II matrices for wide gamut monitors?
Post by: annamaerz on June 22, 2010, 03:26:40 pm
Quote from: digitaldog
As far as I know, the DTP94 was not optimized for wide gamut displays.

Is it difficult, expensive to do?
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 22, 2010, 03:30:49 pm
Quote from: annamaerz
Is it difficult, expensive to do?

Considering its a unsupported product, no longer manufactured, yes.
Title: SpectraView II matrices for wide gamut monitors?
Post by: Paul Sumi on June 22, 2010, 03:45:13 pm
Quote from: digitaldog
As far as I know, the DTP94 was not optimized for wide gamut displays.

You're correct, it is not.  I'll eventually replace mine, but my DTP94 does a pretty good job with SVII (Windows version) and my 2690.  But I'm not super-critical, so YMMV.

Paul
Title: SpectraView II matrices for wide gamut monitors?
Post by: dctech on June 23, 2010, 03:27:45 am
Hello,

our color sensors from Spyder3 family (which is the 3rd generation) are compatible with wide gamut displays.
Please keep in mind that there is no reason to be suspicious of the Spyder3 because we are offering a 30 day money back guarantee.
So just give it a try.

Best regards,
Oliver Mews
Switzerland
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 03:57:42 am
Quote from: dctech
Please keep in mind that there is no reason to be suspicious of the Spyder3 because we are offering a 30 day money back guarantee.
So just give it a try.

Let's assume, after reading this, some novice, would order a spyder3, and he knows that, even though he can't tell the difference right now, he will most probably have learned enough in another two months, to be able to fully assess the instruments capabilities. But right now he doesn't know what to look for. What test should he run?

Quote from: dctech
no reason to be suspicious of the Spyder3

according to the report mentioned earlier there is maybe even reason to be a tiny little bit less suspicious about a spyder3 than about a regular i1d2, right?

Quote
Inter instrument agreement between at least 20 units:

- EyeOne Display 2 - max 18E, mean 8 E
- Spyder 3 Elite - max 15 E, mean 7 E
- DTP94b - max 3 E, mean 1,5E
- EyeOne Pro Rev D - max 3 E, mean 2 E

Though inter instrument agreement shouldn't mean anything if you only use one, or not?
Title: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on June 23, 2010, 05:14:08 am
Quote from: shayaweiss
according to the report mentioned earlier there is maybe even reason to be a tiny little bit less suspicious about a spyder3 than about a regular i1d2, right?

In case of PA241W I wouldn't worry - it works flawlessly even without a calibration sensor. I'd bet it's better calibrated out of the box, than after treating it with some Spyder3 or i1d2.
Title: SpectraView II matrices for wide gamut monitors?
Post by: AndreaPiaggesi on June 23, 2010, 06:57:29 am
Thank you very much for all the information but the colirmeter choice now is even more complicated!! So we've added the DTP94 to the list

- i1Display 2
- Spyder3
- DTP94

Actually the Quato company is selling the DTP94 for about 200 EUR here in Europe (they call it Quato Silver Haze), it's a bit expensive compared with the i1d2 (about EUR 130) and the Spyder3 (about 90 EUR).

I don't know if Quato is selling the standard DTP94 or a modified version of it, but they recommed this package to calibrate their really expensive wide-gamut displays.

Probably an amatour photographer doesn't need a perfect calibration of a good monitor like the PA241W (I'm sure that anybody will get decent result with the default settings) but now I'm really confused reagarding all these colirmeters.

Does anybody know a USA reseller that have the NEC sensor in stock (other than B&H)?
Thanks!!
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 23, 2010, 08:00:15 am
Quote from: shayaweiss
I don't need decades of experience to see that a monitor has color cast or doesn't much a second monitor. I do also not need that kind of background, to see that a change of Instrument and/or software does change the outcome, and that's no anticipatory talk.

For the rest, as I told you, your critics are pointed at the wrong person, I am not a specialist and have no experience. You say different then the others (I gave you a whole list already), I don't mind, you still might be right, but chances are slim. That is what my professional experience is able to tell, as I spent decades in humanities. For the rest I am just not the right interlocutor.

I would be very pleased to follow your exchange with them. As my reason for writing here is to understand and learn. If you want to knock me down with years of experience, even though what you say goes against this or that view expressed here and there by others with decades of experience, than there is no reason to talk anymore.

Again, I would be pleased if these technical issues could be addressed in a civilized manner from now on.

thank you.

P.S.: Very interesting that you do not address your points with Czornyj, whom I was quoting among others.
This is getting silly.  You don't even have a NEC monitor yet you're basing your entire argument as if you have one right in front of you just because of what you're read on the internet.  Lets be very clear, the slight differences these colorimeters make when used on an NEC system will not produce a "color cast."  A slightly 'off' whitepoint does not a 'color cast' make.

I'm not saying I am different than anyone.  I'm not saying anyone is wrong.  What I'm saying.. more than a few times.. is that my i1d2 colorimeter makes a perfect to my reference color checker and it does it across both monitors.  And I think it's more than silly to say NEC is approving colorimeters for their monitors and then telling its customers this one works great, this one works 'not so good', etc.. Either they meet their minimum specifications or they do not.

You're making several huge mistakes on this issue.

1.  You're assuming the colorimeters will behave the same with NEC monitors as they have for your Eizo.  Clearly, if you've read the material you posted.. you'd know they do not.

2.  You're basing the defect of a "color cast" to a slightly inaccurate whitepoint.  There are more than a few degrees between the two.

And you're being especially silly to think I'm going to have exchanges with anyone who wrote anything on the internet not directly addressing me in this thread.  And yes, I've posted to Czornyi on several occasions and he's answered my questions on several occasions.

And do stop being silly about this not being civilized.  We call that whining.  No one has been rude to you or done anything but be very generous in taking time to help you.  I told you when this started that the reason people don't respond to questions in this area is because the person asking the question usually isn't ready to be helped.  They do exactly what you've done, they start many threads asking basically the same questions over and over again in different ways until they get the answer they 'think' they wanted in the first place.

To be blunt you're what I call a "Nervous Nelly" who wants all the answers set in stone up front.  This isn't that type of subject..
Title: SpectraView II matrices for wide gamut monitors?
Post by: Mark Paulson on June 23, 2010, 08:38:35 am
Quote from: Steve Weldon
This is getting silly.  You don't even have a NEC monitor yet you're basing your entire argument as if you have one right in front of you just because of what you're read on the internet.  Lets be very clear, the slight differences these colorimeters make when used on an NEC system will not produce a "color cast."  A slightly 'off' whitepoint does not a 'color cast' make.

I'm not saying I am different than anyone.  I'm not saying anyone is wrong.  What I'm saying.. more than a few times.. is that my i1d2 colorimeter makes a perfect to my reference color checker and it does it across both monitors.  And I think it's more than silly to say NEC is approving colorimeters for their monitors and then telling its customers this one works great, this one works 'not so good', etc.. Either they meet their minimum specifications or they do not.

You're making several huge mistakes on this issue.

1.  You're assuming the colorimeters will behave the same with NEC monitors as they have for your Eizo.  Clearly, if you've read the material you posted.. you'd know they do not.

2.  You're basing the defect of a "color cast" to a slightly inaccurate whitepoint.  There are more than a few degrees between the two.

And you're being especially silly to think I'm going to have exchanges with anyone who wrote anything on the internet not directly addressing me in this thread.  And yes, I've posted to Czornyi on several occasions and he's answered my questions on several occasions.

And do stop being silly about this not being civilized.  We call that whining.  No one has been rude to you or done anything but be very generous in taking time to help you.  I told you when this started that the reason people don't respond to questions in this area is because the person asking the question usually isn't ready to be helped.  They do exactly what you've done, they start many threads asking basically the same questions over and over again in different ways until they get the answer they 'think' they wanted in the first place.

To be blunt you're what I call a "Nervous Nelly" who wants all the answers set in stone up front.  This isn't that type of subject..

ROFLMAO. The lady has made so may post on the same subject that I'm confused! I listened to two people who make a living doing color consulting and both told me to ditch the DTP 94, buy the Spyder 3 to use with my Eizo CG 222w. A long time ago they said use Color Eyes for the software and I have never looked back. I also have a i1Pro UV cut and both guys said that would work with the Eizo, but the Spyder would be better. They were right. I also just recently received my new Solux Lights. What a difference. I wanted 3700K bulbs, but had to settle for the 3500K new screw in type. The color cast in that room is gone. As Andy has said many times...quite worrying about the numbers and view the prints.
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 23, 2010, 09:20:51 am
Quote from: Czornyj
In case of PA241W I wouldn't worry - it works flawlessly even without a calibration sensor. I'd bet it's better calibrated out of the box, than after treating it with some Spyder3 or i1d2.

I suspect so when you consider the cost of the spectroradiometer used by NEC at the factory to set up the calibration. Its impressive that we have a display that you don’t have to calibrate although NEC allows this if you feel you want to. I would like to run trending reports over say a year and see just how well this all works. But on paper, its a giant leap forward for users.
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 23, 2010, 09:21:46 am
Quote from: AndreaPiaggesi
Thank you very much for all the information but the colirmeter choice now is even more complicated!! So we've added the DTP94 to the list

Take it off the list. Its no longer manufactured.
Title: SpectraView II matrices for wide gamut monitors?
Post by: annamaerz on June 23, 2010, 09:28:37 am
Quote from: Steve Weldon
This is getting silly.

IMHO you are being extremely rude and condescending. I followed this thread and am wondering whether your are not able to listen or read. Please stop the pollution now, both.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 09:34:41 am
Quote from: digitaldog
I suspect so when you consider the cost of the spectroradiometer used by NEC at the factory to set up the calibration. Its impressive that we have a display that you don’t have to calibrate although NEC allows this if you feel you want to. I would like to run trending reports over say a year and see just how well this all works. But on paper, its a giant leap forward for users.

How does this technique actually work? How can it correct over time the drift? This must surely vary from monitor to monitor? Can you point me to some reference? What is it called?

thanks

shaya
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 23, 2010, 09:39:22 am
The display is factory calibrated using a very expensive lab grade spectroradiometer. It is then capable of tracking how colors change over time and temperature etc. An internal color engine allows you to dial in a specific luminance, white point, gamut etc. and it will calibrate to the target aim point without the need for an external sensor.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 09:49:13 am
Quote from: digitaldog
The display is factory calibrated using a very expensive lab grade spectroradiometer. It is then capable of tracking how colors change over time and temperature etc. An internal color engine allows you to dial in a specific luminance, white point, gamut etc. and it will calibrate to the target aim point without the need for an external sensor.

Why should one still want an instrument? Is it still that little bit more precision that you get, if you need it?
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 23, 2010, 09:51:34 am
Quote from: shayaweiss
Why should one still want an instrument? Is it still that little bit more precision that you get, if you need it?

My understanding is the instrument can be used to update the internal calibration process over time for the ultimate precision.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 10:03:33 am
Quote from: digitaldog
My understanding is the instrument can be used to update the internal calibration process over time for the ultimate precision.

I see the software is called MultiProfiler and controls the SpectraView Engine, which does this automatic tracking ( http://www.necdisplay.com/MultiProfiler/ (http://www.necdisplay.com/MultiProfiler/) ).

It allows for internal "resetting/recalibration" of this self-calibration process?:

Quote
QUESTION: Is it possible to calibrate the monitor using an external color sensor?

ANSWER: Yes. The SpectraView software is available to calibrate the monitor if necessary. In addition, the OSD or MultiProfiler controls can be used together with a 3rd party application to manually adjust the monitor together with an external color sensors if necessary.

from FAQ ( http://www.necdisplay.com/MultiProfiler/faq/ (http://www.necdisplay.com/MultiProfiler/faq/) )
Title: SpectraView II matrices for wide gamut monitors?
Post by: WillH on June 23, 2010, 10:18:16 am
Quote from: shayaweiss
I see the software is called MultiProfiler and controls the SpectraView Engine, which does this automatic tracking ( http://www.necdisplay.com/MultiProfiler/ (http://www.necdisplay.com/MultiProfiler/) ).

It allows for internal "resetting/recalibration" of this self-calibration process?

A future version of SpectraView will allow you to use your sensor values to update the internal factory measurements if desired, but this will really only have an impact when used together with MultiProfiler. The reason being that SpectraView calibrates to the sensor values each time anyway. MultiProfiler uses the very accurate factory measurements for each individual display, combined with internal monitoring to achieve very good results without the need for an external sensor or calibration package.
Title: SpectraView II matrices for wide gamut monitors?
Post by: AndreaPiaggesi on June 23, 2010, 10:20:21 am
Quote from: digitaldog
Take it off the list. Its no longer manufactured.


Sorry but, as I've wrote previously, Quato (a German company that produce high end display) is actually selling the DTP94 with its calibration software. It's called iColor.
http://www.amazon.de/gp/product/B002UZ5F3K...me=&seller= (http://www.amazon.de/gp/product/B002UZ5F3K/ref=olp_product_details?ie=UTF8&me=&seller=)

Here's an interesting Whitepaper:
http://www.quato.de/german/produkte/Whitepaper_IP_E.pdf (http://www.quato.de/german/produkte/Whitepaper_IP_E.pdf)

This is just a snippet:
Colorimeter support for Wide Gamut units
A wide color gamut unit forces the user to have an expensive and limited precision Spectrophotometer, but Quato has found a way to make use of the highly precise DTP94 colorimeter by uploading a correction table into the device. Now users can benefit from the much higher precision of the colorimeter (compared to the Spectro) and save a lot of money.

Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 10:22:11 am
Quote from: WillH
A future version of SpectraView will allow you to use your sensor values to update the internal factory measurements if desired, but this will really only have an impact when used together with MultiProfiler. The reason being that SpectraView calibrates to the sensor values each time anyway. MultiProfiler uses the very accurate factory measurements for each individual display, combined with internal monitoring to achieve very good results without the need for an external sensor or calibration package.

Thank you for this information.

Is there a time line set for this update of SpectraView, the software?
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 23, 2010, 10:50:22 am
Quote from: annamaerz
IMHO you are being extremely rude and condescending. I followed this thread and am wondering whether your are not able to listen or read. Please stop the pollution now, both.
Being rude and condescending certainly wasn't my intent, no more than it was your intent by telling someone they can't read or listen.  Like you I was doing my very best to get a message across, to help someone.  I failed.  Of that there can be no misunderstanding.
Title: SpectraView II matrices for wide gamut monitors?
Post by: nilo on June 23, 2010, 11:13:52 am
Quote from: AndreaPiaggesi
Sorry but, as I've wrote previously, Quato (a German company that produce high end display) is actually selling the DTP94 with its calibration software. It's called iColor.
http://www.amazon.de/gp/product/B002UZ5F3K...me=&seller= (http://www.amazon.de/gp/product/B002UZ5F3K/ref=olp_product_details?ie=UTF8&me=&seller=)

Here's an interesting Whitepaper:
http://www.quato.de/german/produkte/Whitepaper_IP_E.pdf (http://www.quato.de/german/produkte/Whitepaper_IP_E.pdf)

This is just a snippet:
Colorimeter support for Wide Gamut units
A wide color gamut unit forces the user to have an expensive and limited precision Spectrophotometer, but Quato has found a way to make use of the highly precise DTP94 colorimeter by uploading a correction table into the device. Now users can benefit from the much higher precision of the colorimeter (compared to the Spectro) and save a lot of money.

In connection to the SpectraView software, I would still think that NEC's own updated, modified or corrected is a better match, than the maybe in general terms better DTP94b!?
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 23, 2010, 11:28:04 am
Quote from: AndreaPiaggesi
Sorry but, as I've wrote previously, Quato (a German company that produce high end display) is actually selling the DTP94 with its calibration software. It's called iColor.
http://www.amazon.de/gp/product/B002UZ5F3K...me=&seller= (http://www.amazon.de/gp/product/B002UZ5F3K/ref=olp_product_details?ie=UTF8&me=&seller=)

I didn’t say it wasn’t being sold, I said its not being manufactured. Its an unsupported X-rite product. Its a great instrument but beware that how much support (think replacement or fixes) you’ll get puts this in a iffy position.
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 23, 2010, 11:28:31 am
Quote from: ninoloss
In connection to the SpectraView software, I would still think that NEC's own updated, modified or corrected is a better match, than the maybe in general terms better DTP94b!?

Yes!
Title: SpectraView II matrices for wide gamut monitors?
Post by: annamaerz on June 23, 2010, 02:06:32 pm
Quote from: Steve Weldon
Being rude and condescending certainly wasn't my intent, no more than it was your intent by telling someone they can't read or listen.  Like you I was doing my very best to get a message across, to help someone.  I failed.  Of that there can be no misunderstanding.

There is a little difference between you and me, you are aggressive and I defend.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 02:46:17 pm
Quote from: digitaldog
Yes!

Also, I don't know if this point is relevant, but the Quato puck is the DTP94 and not the DTP94b, which, if I am not mistaken is the acclaimed one. I have a DTP94b and it does indeed perform marvelously well on all our standard monitors. Even on the very cheap LG227w wide gamut monitor, when used with software correction (Quato), or on the laptops here, which I had to recalibrate today. It was sold to me as "EIZO DTP94b", though it does not say EIZO anywhere on it. And it looks different than the DTP94.
Title: SpectraView II matrices for wide gamut monitors?
Post by: nilo on June 23, 2010, 03:55:14 pm
After everything that was discussed and quoted about correction filters and matrices for wide gamut monitors - and only one person disagreeing without real arguments -, I don't understand how  it is possible that x-rite does not incorporate such an optional correction in their software?
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 23, 2010, 03:59:20 pm
Quote from: ninoloss
After everything that was discussed and quoted about correction filters and matrices for wide gamut monitors - and only one person disagreeing without real arguments -, I don't understand how  it is possible that x-rite does not incorporate such an optional correction in their software?

You may mean hardware. That’s really where these filter matrices need to live.
Title: SpectraView II matrices for wide gamut monitors?
Post by: annamaerz on June 23, 2010, 03:59:28 pm
Quote from: ninoloss
After everything that was discussed and quoted about correction filters and matrices for wide gamut monitors - and only one person disagreeing without real arguments -, I don't understand how  it is possible that x-rite does not incorporate such an optional correction in their software?

And with iColor with do see it's feasible...
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 04:05:00 pm
Quote from: annamaerz
And with iColor with do see it's feasible...

When asked the same question, about why not integrate a correction into Argyll, Graeme Gill answerd the following:
Quote
[argyllcms] Re: Correction for wide gamut screens and/or RGB-LED?

    * From: Graeme Gill <graeme@xxxxxxxxxxxxx>
    * To: argyllcms@xxxxxxxxxxxxx
    * Date: Tue, 06 Apr 2010 09:14:43 +1000

Steffen wrote:

  >  How does ArgyllCMS cope with this, are there any plans (or even just
  > data on the deviation) for the future? I find that not only software

Currently the instruments are assumed to provide correct XYZ values.
Generally you get what you pay for in this regard. Some colorimeters
don't work so well on some displays, and this is a good reason for
buying a spectrometer.

While it's perfectly possible to create correction matrices for particular
display/colorimeter combinations, in practice there are stumbling blocks.

A major one is that it's a combinatorial explosion. I don't even have
access to all the displays that support in-display Luts, never mind the
thousands
of other LCD displays out there, so I can't supply such correction
matrices. (I suspect it would be a full time job keeping up with them,
even if one were suppled monitors by all the manufacturers). The other
problem is that one particular instance of a colorimeter doesn't necessarily
represent the average for that type of instrument. One needs access to a
statistically significant number of instruments to overcome this problem.

[A systematized solution would involve the display manufacturers publishing
 the spectral characteristics of their displays in a standard format, and
 having some means of measuring the spectral sensitivities of the instruments.]


 >   developers have to face the problem of designing their software so that
  >  it uses some form of CMM, but that the profiling process itself is
  >  affected by the changes in display technology.

The profiling is not affected. Shortcuts taken by the colorimeter makers
do affect their accuracy though.

Graeme Gill.

source: http://www.freelists.org/post/argyllcms/Co...-andor-RGBLED,1 (http://www.freelists.org/post/argyllcms/Correction-for-wide-gamut-screens-andor-RGBLED,1)
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 23, 2010, 04:06:18 pm
Quote from: shayaweiss
When asked the same question, about why not integrate a correction into Argyll, Graeme Gill answerd the following:

Which makes you wonder about iColor no?
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 04:07:47 pm
Quote from: digitaldog
You may mean hardware. That’s really where these filter matrices need to live.

You mean like NEC does with the i1d2 puck?

Would it disturb when used with a non wide gamut monitor?
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 04:08:29 pm
Quote from: digitaldog
Which makes you wonder about iColor no?

Exactly, that was my next silly question.
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 23, 2010, 04:10:32 pm
Quote from: shayaweiss
You mean like NEC does with the i1d2 puck?

But in this case, the filters ARE mated to the display. Its not some correction happening in software, based on some assumption of the display(s). If not, why would NEC go to the trouble if they could just pop this into SpectraView software. Not cheap having a pile of custom built EyeOne Displays OEM’ed. FWIW, if you look back at Barco, PressView and Artisan, they too used custom mated filters for old, sRGB like CRTs. They did it for a reason.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 04:33:18 pm
Quote from: digitaldog
But in this case, the filters ARE mated to the display. Its not some correction happening in software, based on some assumption of the display(s). If not, why would NEC go to the trouble if they could just pop this into SpectraView software. Not cheap having a pile of custom built EyeOne Displays OEM’ed. FWIW, if you look back at Barco, PressView and Artisan, they too used custom mated filters for old, sRGB like CRTs. They did it for a reason.

trying to sum up:

So, with wide gamut monitors, one should only use the custom mated puck (or a spectrometer, although some people don't recommend the latter for displays). According to what we came up with here,  if one needs calibration with a device for a wide gamut monitor, one should use a monitor for which there is such a custom mated puck. Also it should be able to be used with the propitiatory software, that can talk to the puck and the monitor.    

In the case of the new NEC PA series, as mentioned, because of the new Profiler software, the need for a puck might be less. As we have heard here, an update of SpectraView is apparently underway. It will supposedly allow recalibration of the self-calibration process of the PA's, via a measurement device, again the NEC corrected puck will be preferable.
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 23, 2010, 04:36:45 pm
Quote from: shayaweiss
So, with wide gamut monitors, one should only use the custom mated puck (or a spectrometer, although some people don't recommend the latter for displays). According to what we came up with here,  if one needs calibration with a device for a wide gamut monitor, one should use a monitor for which there is such a custom mated puck. Also it should be able to be used with the propitiatory software, that can talk to the puck and the monitor.


You don’t have to but if you have no instrument, by all means if given the opportunity, get a mated colorimeter.

You can buy a suit off the rack or have one custom tailored. Its not like the suit off the rack is akin to walking the streets naked.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 04:42:01 pm
Quote from: digitaldog
You don’t have to

let me seize this opportunity to try to clear something up, which was a matter of wonder for some of us:

We concluded, just now, that a custom mated puck is preferable, right?

How can we understand that NEC lists almost all existing colorimeters, which are not corrected, as compatible with SpectraViewII and wide gamut monitors?

The only explanation, which is very weak, I agree with Steve Weldon, is that, like you said you don't have to, but its better to get the custom tailored one, if you can?!

I am happy to see that you have the same explanation.

Anyone?
Title: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on June 23, 2010, 05:23:50 pm
Quote from: shayaweiss
let me seize this opportunity to try to clear something up, which was a matter of wonder for some of us:

We concluded, just now, that a custom mated puck is preferable, right?

How can we understand that NEC lists almost all existing colorimeters, which are not corrected, as compatible with SpectraViewII and wide gamut monitors?

The only explanation, which is very weak, I agree with Steve Weldon, is that, like you said you don't have to, but its better to get the custom tailored one, if you can?!

I am happy to see that you have the same explanation.

Anyone?

Like Graeme said - "The profiling is not affected. Shortcuts taken by the colorimeter makers do affect their accuracy though" - so, as a matter of fact, they're all compatible, it's only a matter that some may be more accurate than others
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 23, 2010, 05:34:25 pm
Quote from: Czornyj
Like Graeme said - "The profiling is not affected. Shortcuts taken by the colorimeter makers do affect their accuracy though" - so, as a matter of fact, they're all compatible, it's only a matter that some may be more accurate than others
Thank you Czornyj.
Title: SpectraView II matrices for wide gamut monitors?
Post by: nilo on June 24, 2010, 05:40:26 am
Quote from: Czornyj
Like Graeme said - "The profiling is not affected. Shortcuts taken by the colorimeter makers do affect their accuracy though" - so, as a matter of fact, they're all compatible, it's only a matter that some may be more accurate than others

I'll add an additional side note to that conclusions:

It puzzles me that, with x-rite's eye-one Match3 software, theres is no way you'll get these more "accurate results" (I think this is only being diplomatic, because we speak in public, but OK   ), as it only might be used with the i1d2 as colorimeter, if I'm not mistaken. As there are no x-rite branded filtered colorimeters for wide gamut, you'll never get it "really" right with Match3, unless, which is your only option, you buy a spectrometer for a few more bucks  . Maybe should even buy two of them, as you'll also want a UV-cut.

For the rest of us, that do not need that "accuracy", meaning "no accuracy at all", they would be maybe better off eyeballing, and honestly that's what I did in the beginning, calibration/profiling to the "approximatly accurat" white point and afterwards adjusting the white point by eye. I did not find any other way of getting acceptable results out of x-rites eye-one match3 + i1d2 + wide gamut monitor.

I didn't try to use the NEC i1d2 with eye match3 (evidently I'm not speaking of using this setup with a hardware calibration NEC monitor)
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 24, 2010, 06:04:16 am
Quote from: ninoloss
I'll add an additional side note to that conclusions:

It puzzles me that, with x-rite's eye-one Match3 software, theres is no way you'll get these more "accurate results" (I think this is only being diplomatic, because we speak in public, but OK   ), as it only might be used with the i1d2 as colorimeter, if I'm not mistaken. As there are no x-rite branded filtered colorimeters for wide gamut, you'll never get it "really" right with Match3, unless, which is your only option, you buy a spectrometer for a few more bucks  . Maybe should even buy two of them, as you'll also want a UV-cut.

For the rest of us, that do not need that "accuracy", meaning "no accuracy at all", they would be maybe better off eyeballing, and honestly that's what I did in the beginning, calibration/profiling to the "approximatly accurat" white point and afterwards adjusting the white point by eye. I did not find any other way of getting acceptable results out of x-rites eye-one match3 + i1d2 + wide gamut monitor.

I didn't try to use the NEC i1d2 with eye match3 (evidently I'm not speaking of using this setup with a hardware calibration NEC monitor)

So after all, if you have no NEC, Color Edge EIZO, Quato or LaCie, it seems, for the "suite of the rack" as Andrew puts it, you are a lot better of with datacolors's spyder3, which is filtered, and a software that accepts it and your monitor?!
Title: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on June 24, 2010, 06:22:56 am
Quote from: shayaweiss
So after all, if you have no NEC, Color Edge EIZO, Quato or LaCie, it seems, for the "suite of the rack" as Andrew puts it, you are a lot better of with datacolors's spyder3, which is filtered, and a software that accepts it and your monitor?!

No. Andrew mentioned about a colorimeter that is optimized for the specific spectra of some NEC displays, and Spyder is supposed to be optimized for anything.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 24, 2010, 06:59:52 am
Quote from: Czornyj
No. Andrew mentioned about a colorimeter that is optimized for the specific spectra of some NEC displays, and Spyder is supposed to be optimized for anything.

Sorry, I don't understand your post? If I understand you, it seams to me that there is no contradiction to what I wrote?!

I think we agree, please do correct me if I am wrong! I meant to write:

If you have NO NEC, NO Color Edge, NO Quato and NO LaCie, so there is NO colorimeter that is specific to your wide gamut device, you will be better of with the spyder3, because the spyder3 is, as Andrew put it, filtered and "optimized for anything".

Maybe you meant?: that as it is "optimized for anything" there is no difference to take a software based correction, like Quato iCOlor. It's equally bad, sorry nino, "less accurate", more "suite of the track"ish.


In advance, please forgive me if got it wrong.
Title: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on June 24, 2010, 07:18:25 am
Quote from: shayaweiss
Sorry, I don't understand your post? If I understand you, it seams to me that there is no contradiction to what I wrote?!

I think we agree, please do correct me if I am wrong! I meant to write:

If you have NO NEC, NO Color Edge, NO Quato and NO LaCie, so there is NO colorimeter that is specific to your wide gamut device, you will be better of with the spyder3, because the spyder3 is, as Andrew put it, filtered and "optimized for anything".

Maybe you meant?: that as it is "optimized for anything" there is no difference to take a software based correction, like Quato iCOlor. It's equally bad, sorry nino, "less accurate", more "suite of the track"ish.


In advance, please forgive me if got it wrong.

I only meant, that I woudn't count there will be a big difference between Eye-One display2 and Spyder3. According to J.Raimar Kuhnen-Burger: "Although, the EyeOne Dispaly 2´s and Spyder 3´s filter may match better to the CIE curves, they are still far off. As a result, all colorimeters need a correction for
wide gamut (and white LED according to our test)"
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 24, 2010, 07:39:45 am
Quote from: Czornyj
I only meant, that I woudn't count there will be a big difference between Eye-One display2 and Spyder3. According to J.Raimar Kuhnen-Burger: "Although, the EyeOne Dispaly 2´s and Spyder 3´s filter may match better to the CIE curves, they are still far off. As a result, all colorimeters need a correction for
wide gamut (and white LED according to our test)"

Noiw I got it, I hope! The regular 1id2 is as off as the spyder3, that's what Kuhnen-Burger says, right?
Title: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on June 24, 2010, 07:54:14 am
Quote from: shayaweiss
Noiw I got it, I hope! The regular 1id2 is as off as the spyder3, that's what Kuhnen-Burger says, right?

Not exactly, he only states that both are "off", and need some correction (not to mention poor inter instrumental agreement).
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 24, 2010, 08:09:56 am
Quote from: Czornyj
Not exactly, he only states that both are "off", and need some correction (not to mention poor inter instrumental agreement).

The regular 1id2 is off. Spyder3 is off. Maybe to a different degree. Do you know of any data, comparisons, test regarding this particular point?!
Title: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on June 24, 2010, 09:11:16 am
Quote from: shayaweiss
The regular 1id2 is off. Spyder3 is off. Maybe to a different degree. Do you know of any data, comparisons, test regarding this particular point?!

http://lists.apple.com/archives/colorsync-...v/msg00204.html (http://lists.apple.com/archives/colorsync-users/2009/Nov/msg00204.html)

Note that these are avareraged measurement data are for selected "best batch" of devices.
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 24, 2010, 09:31:43 am
Quote from: annamaerz
There is a little difference between you and me, you are aggressive and I defend.
There is certainly a difference between you and me, but it is hardly your place to publicly state this difference any more than it is mine to give my opinion of why you're choosing to get involved in an exchange that was not directed towards you.  And yes I do have an opinion, I just don't feel an overwhelming need to let you know what it is.  Perhaps that is one difference?

Sure, I could come up with something "rude and condescending" like "I wonder if you can listen or read", (words inside quotes are your words) but really what purpose would it serve other than to elevate the discourse?

I am direct and I'll continue to be direct on this subject because I think it needs to be said:  It doesn't seem to be enough that someone asks for and receives help.  They also demand the help arrives in a manner in which they approve.  I apologize if "you're being silly" is seen as some sort of terrible insult.  Believe me, it wasn't my first choice of words.  I had to seriously search for the most polite term I could muster.  However it was taken, it doesn't make it any less accurate.

When I ask for help I'm willing to overlook relatively innocent terms that may result when one of the parties is using English as a second or third language, as well as inflections from the same.  There is also the all too present fact that written words without human exchange are frequently taken differently than intended.  Again, language exacerbates this phenomenon.   The truth is, electronic discussion is fraught with potential misunderstandings in attitude, intent, and so forth and language as well as regional differences can make it much worse.  I can overlook these sort of things, even from a "defender", because I'm more interested in the subject matter than protecting my own fragile feelings.

Not everyone will agree and it would be 'silly' of me to expect everyone to agree.  Another reason why concentrating only on the subject is more productive than rendering intentional insults.

 

Title: SpectraView II matrices for wide gamut monitors?
Post by: nilo on June 24, 2010, 10:07:49 am
Quote from: Czornyj
http://lists.apple.com/archives/colorsync-...v/msg00204.html (http://lists.apple.com/archives/colorsync-users/2009/Nov/msg00204.html)

Note that these are avareraged measurement data are for selected "best batch" of devices.
Title: SpectraView II matrices for wide gamut monitors?
Post by: annamaerz on June 24, 2010, 10:15:39 am
Quote from: Steve Weldon
I apologize if "you're being silly" is seen as some sort of terrible insult.  Believe me, it wasn't my first choice of words.  I had to seriously search for the most polite term I could muster.  However it was taken, it doesn't make it any less accurate.

Hmm. You can't help it?
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 24, 2010, 10:53:04 am
Quote from: Czornyj
http://lists.apple.com/archives/colorsync-...v/msg00204.html (http://lists.apple.com/archives/colorsync-users/2009/Nov/msg00204.html)

Note that these are avareraged measurement data are for selected "best batch" of devices.

Thank for that link.
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 24, 2010, 01:04:10 pm
Quote from: annamaerz
Hmm. You can't help it?
LOL.. fair question.  I suppose like most others it depends.  It's an area I can work on..
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on June 27, 2010, 01:50:40 am
Quote from: Steve Weldon
NEC specifically told me the i1d2 supports wide gamut and its listed in their manual as doing so..

Not only that.. it actually works and profiles the wide gamut as you can clearly see from the attachments..

After you buy your NEC puck.. it will be interesting to see you run it side by side to your current i1d2 puck and show us the difference..

Still curious about the differences of the Euro vs. America vs. Asian spec monitors..

[attachment=22722:i1d2.JPG]  [attachment=22723:i1d2a.JPG]

It does seem pretty tricky to figure out exactly the difference between the NEC puck and a stock ID2 is. It seems to me that they must be using at least a bit of a translation matrix for their puck or have altered the filters or something. I regular ID2 using regular software gives weird results on wide gamut monitors. Stock ID2 pucks also seem to vary quite a bit puck to puck, hopefully the NEC ones have all been calibrated to some specific matching standard.

As a quick test I used two DTP94 pucks and the NEC puck on a standard sRGB-ish gamut monitor and compared the results.

All three gave fairly similar gamma ramp shapes. Really almost exactly the same from 10%-90%  (above 90% the NEC puck started giving a bit different shape).

The NEC puck claimed the black point was deeper than the DTP94 pucks did. On an HTDV both DTP94s put the blackpoint at 0.020 (with white point near 100 cd/m^2) while the NEC puck put the blackpoint at 0.015 and the WP 2cd/m^2 higher and registered a significantly higher claimed contrast ratio.


However, when it came to reading the location of the primaries the NEC puck started to read things differently and when it came to R,G,B balance on the gray-scale ramp it differed considerably from the DTP94 (which both gave pretty similar readings). It seemed to place green at a higher x-coord than the DTP94s and red at a lower x-coord and blue at a higher luminance.

On the set as is the NEC placed both red and green dead on but had blue 5-10% weak along much of the way while both DTP had instead blue and green almost aligned with another and just a touch weak but red an even 10% too hot compared to B/G.  So NEC had green and red aligned and blue trailing and the DTPs had green and blue aligned with red leading. Not really the greatest agreement there between the NEC and either DTP94.

So I'm rather hoping that the NEC puck did have something special done to it that makes it no longer work ideally with alternate software on sRGB monitors, otherwise it's lack of agreement when it came to channel balance and primary readings with the DTP94s is perhaps a bit disconcerting.

One very odd thing is that a program that used compensation matrices had the two DTP94s giving somewhat different results in terms of WB on a wide gamut monitor, even though using regular software on an sRGB monitor they gave pretty similar results.

One of the DTP94 in the generic IPS wide gamut DTP94 mode reported 6450 WB (after calibrated with the NEC puck using spectraview II), the NEC puck gave more like 6670WB though with that software. Perhaps this hints that the NEC puck works at least somewhat well with Spectraview II and yet not as well using other software and thus has had something done to it to make it act differently than a stock ID2?

Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on June 27, 2010, 02:00:55 am
Quote from: Czornyj
European Spectraview display profiler has correction matrix for NEC display spectra and popular colorimeters, so it's the same approach as Eizo ColorNavigator, and Quato iColor display.
NEC Display America choosed to build custom mated colorimeter rather than make software matrix correction. The Spectraview II checks the sensor type, and when it's not NEC colorimeter or spectrophotometer it creates profiles using factory avaraged measurements taken during production. See Spectraview II Edit>Preferences>ICC Profile tag.

Testing the profile quality against CC24 gives only overall idea of how the profile works, and as a matter of fact it's impossible to get 100% visual connection - the spectra of the displays backlight and the spectra of our lighting is different, and so on...

Interesting, so the European spectraview have tables for say DTP94s?

(although apparently they require such a large correction that they lose a fair amount of precision afterwards.... although then why would quato do this?)
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on June 27, 2010, 02:09:24 am
Quote from: digitaldog
Take it off the list. Its no longer manufactured.

yes, it is, they started production again.

you can only get it through one vendor now though in the US (perhaps also Quato in Europe???).
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 27, 2010, 10:27:31 am
Quote from: LarryBaum
yes, it is, they started production again.

you can only get it through one vendor now though in the US (perhaps also Quato in Europe???).

X-Rite started up production? You sure? OR you’re saying a stock of built units is still available? Big difference.
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 27, 2010, 11:54:11 am
Quote from: LarryBaum
It does seem pretty tricky to figure out exactly the difference between the NEC puck and a stock ID2 is.
From what we've learned here.. There IS a difference between pucks.  There is also a difference in the software.. some software having built in profiles for supported pucks (NEC for instance), and others don't have these profiles.  Without the profiles I'd image the variation would be quite pronounced.. as in Shaya using the 1di2 with her Eizo monitors.. while me using the same puck with my SVII software (that has built in profiles to support the 1id2 puck) has the potential to be spot on (depending on how close my individual puck comes to their 'average' sample, and the accuracy of the profile), or somewhat of.

From experience I can tell you it appears to be "spot on."  And by "spot on" this means withing a few hundred degrees that's easy enough to adjust/compensate for in your printer profile or whatever else your use 'might' be.  

I have no illusions a 1id2 puck, using SVII software, on ANY NEC monitor will produce noticeable color casts.  It won't.   The same puck, used with Imatch or some other software without the profiles to support a non-wide gamut puck to a wide gamut monitor.. potentially WILL cause color casts.

So.. in the interest of "I'm just damn curious" I ordered a SVII puck through Amazon USA, delivered here to me in Thailand for $318 and $35 shipping (should receive it before the 20th of July).. and we'll see if there are any earth shattering differences.  

AND we'll find out of if the NEC SVII puck calibrated for wide gamut monitors.. works on ALL wide gamut monitors such as the Eizo.

Hell, my displays appear to be and are functioning as perfect.. but in the interest of science (I say that with a half-way straight face) I'm willing to spend a few bucks to find out.


This leads me to my next issues I'm experiencing.  In sRGB emulation mode.. when I process an image to 'average' levels every monitor out there loves it.. but when I put the luminance, contrast, and color so they'll look great on my NEC monitor in sRGB emulation mode.. it appears uncalibrated laptops and cheaper desktop displays without sRGB emulation.. display these files poorly.. to the point they obviously look wrong.

SO.. we'll need a range of adjustment for sRGB "everyone" and then sRGB for those with a full set of calibrated equipment with including sRGB Emulated monitors..

This means.. we process to our audience..  Wonderful unless you don't know who you audience is.

Nail in the head stuff..

But the fact is.. without the monitors and pucks and software in hand.. trying to figure out what will happen "for sure" is (can't say silly, it offends) so.. is.. well.. less than idea..

And of course so far we're only taking about monitors and not the rest of the color system like scanners, printers, publishers, print houses, etc..  And I don't even want to get into the story where I recently calibrated/profiled (by eye because it was all I had.. I knew what my files 'should' look like) a major print houses gear.. and had the owner come out and tell me "that's the best our prints have even looked ever..   These are 12-14 meter long laser printers with four print heads spitting out from 4x6's to 5x7's t 8x10's to 11x14's to 20x24's. Huge investments in gear.. and not a puck or software in site.  It was literally fly by wire.. and they were estatic.  I handed them a card and told them if they made an appointment I'd bring in the right gear and get them set up.  They seemed eager until I told them these calibrations should really be done weekly if not daily in a big print house..

And so the world turns..   So far I have three pucks.. nothing suitable.  Soon I'll have four.  Will the NEC SVII be the major one?
Title: SpectraView II matrices for wide gamut monitors?
Post by: probep on June 27, 2010, 02:20:32 pm
Quote from: Steve Weldon
From what we've learned here.. There IS a difference between pucks.
Yes, take a look at http://luminous-landscape.com/forum/index....mp;#entry351434 (http://luminous-landscape.com/forum/index.php?showtopic=42040&st=0&p=351434&#entry351434)
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 27, 2010, 02:54:59 pm
Quote from: probep
Yes, take a look at http://luminous-landscape.com/forum/index....mp;#entry351434 (http://luminous-landscape.com/forum/index.php?showtopic=42040&st=0&p=351434&#entry351434)
Sure.. but you know.. 25 Corvette ZR1's coming off the product line one after the other, 25 F200/2 IS Canon lenses, 100 rolls of 4x5 Velvia film.. no matter what it is. there will be a variation of manufacturing tolerances which may or may not affect our personal use of the item depending on how we use that item and what for..

I personally think people are concentrating FAR too much on how the screen looks compared to a theoretical standard.  In our minds we all want it to be perfect.. but include this into the rest of our color workflow, printer profiles, variations in ambient light during different times of the day, how tired our eyes our, and so on and so on.. and these small differences mean less and less.

However, like everyone else I'd love to have perfect color.  But perfect color ONLY exists in my personal closed workflow system.. and perhaps extends to my print house (which I have no control over) or a publisher (even less control).. meanwhile the other 99% of the world looking at our images on the web are using 4-5 year old laptops with faded LCD's not even cable of viewing 30-40% of the sRGB gamut..

Really... I think "silly" fit.. no offense intended..
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on June 27, 2010, 03:37:30 pm
Quote from: digitaldog
X-Rite started up production? You sure? OR you’re saying a stock of built units is still available? Big difference.

well spectracal claims that they actually started up production again

if they are telling the truth, then yes
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on June 27, 2010, 03:54:11 pm
Quote from: probep
Yes, take a look at http://luminous-landscape.com/forum/index....mp;#entry351434 (http://luminous-landscape.com/forum/index.php?showtopic=42040&st=0&p=351434&#entry351434)

Does you NEC puck actually say "MDSVSENSOR2" on it or just "MDSVSENSOR". I ordered mine separately through Amazon and it only has the name without the "2" stamped on top. Is that old stock? (although calman appears to be able to read internal data out of the puck and says that it was calibrated this April, which doesn't sound like old stock at all)   Maybe the 2 just refers to the software version spectraview "II".

Title: SpectraView II matrices for wide gamut monitors?
Post by: probep on June 27, 2010, 04:53:31 pm
Quote from: LarryBaum
Does you NEC puck actually say "MDSVSENSOR2" on it or just "MDSVSENSOR". I ordered mine separately through Amazon and it only has the name without the "2" stamped on top. Is that old stock?
I've bought two NEC MDSVSENSOR2 colorimeters at NEC Online Store (http://www.necdisplay.com/Products/Accessories/?product=8899a96d-28dc-484f-a4de-14309a636738): one in January and another in May 2010. Each is with "MDSVSENSOR" label and looks like that (http://www.flickr.com/photos/miahz/3989679014/).
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on June 27, 2010, 07:17:52 pm
Quote from: probep
I've bought two NEC MDSVSENSOR2 colorimeters at NEC Online Store (http://www.necdisplay.com/Products/Accessories/?product=8899a96d-28dc-484f-a4de-14309a636738): one in January and another in May 2010. Each is with "MDSVSENSOR" label and looks like that (http://www.flickr.com/photos/miahz/3989679014/).

Thanks.

Yeah, that looks exactly the same as mine. I guess they don't label them as "2".

(and with calman reporting a mid-April 2010 calibration date, safe to say the one I got is the current puck)

I just wanted to be sure that maybe there were not some left over spectraview I kits and that maybe they used a different puck (perhaps adapted to the smaller wide gamuts of some of the older models).
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 28, 2010, 01:30:16 am
Quote from: LarryBaum
Does you NEC puck actually say "MDSVSENSOR2" on it or just "MDSVSENSOR". I ordered mine separately through Amazon and it only has the name without the "2" stamped on top. Is that old stock? (although calman appears to be able to read internal data out of the puck and says that it was calibrated this April, which doesn't sound like old stock at all)   Maybe the 2 just refers to the software version spectraview "II".
Larry -  If you wouldn't mind terribly..  Would you give me the link on Amazon where you ordered yours?  Mine hasn't shipped yet and it would be somewhat reassuring they're sending me the right one.  There were two places that claim to have them.. The Amazon Market Place (the only place that will ship to me here overseas) and a place called Proline.

It would be somewhat irritating to pay the shipping over here, not to mention the heavy VAT and customs fees.. and get an old one..

Thanks!
Title: SpectraView II matrices for wide gamut monitors?
Post by: AndreaPiaggesi on June 28, 2010, 08:34:59 am
Quote from: Steve Weldon
Sure.. but you know.. 25 Corvette ZR1's coming off the product line one after the other, 25 F200/2 IS Canon lenses, 100 rolls of 4x5 Velvia film.. no matter what it is. there will be a variation of manufacturing tolerances which may or may not affect our personal use of the item depending on how we use that item and what for..

I personally think people are concentrating FAR too much on how the screen looks compared to a theoretical standard.  In our minds we all want it to be perfect.. but include this into the rest of our color workflow, printer profiles, variations in ambient light during different times of the day, how tired our eyes our, and so on and so on.. and these small differences mean less and less.

However, like everyone else I'd love to have perfect color.  But perfect color ONLY exists in my personal closed workflow system.. and perhaps extends to my print house (which I have no control over) or a publisher (even less control).. meanwhile the other 99% of the world looking at our images on the web are using 4-5 year old laptops with faded LCD's not even cable of viewing 30-40% of the sRGB gamut..

Really... I think "silly" fit.. no offense intended..

I completely agree with you but I usually spend a lot of time in RAW conversion, rating, adjustments, Photoshop editing etc ... and probably a perfect monitor with theoretical perfect colors will avoid strange color cast in photolab's prints and will help to get a correct file for your archive.
Tomorrow when you'll buy the next monitor your pictures will look almost the same, without big color shift.

Anyway I agree with you, nobody will see our perfect colors, they'll be limited to our workflow.


Quote from: LarryBaum
Does you NEC puck actually say "MDSVSENSOR2" on it or just "MDSVSENSOR". I ordered mine separately through Amazon and it only has the name without the "2" stamped on top. Is that old stock? (although calman appears to be able to read internal data out of the puck and says that it was calibrated this April, which doesn't sound like old stock at all)   Maybe the 2 just refers to the software version spectraview "II".


Maybe I'm wrong but the MDVSENSOR2 is the custom calibrated sensor especially made for wide-gamut LCD; the MDVSENSOR is the previous version.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 28, 2010, 08:49:01 am
Quote from: AndreaPiaggesi
Anyway I agree with you, nobody will see our perfect colors, they'll be limited to our workflow.

I might be a rare exception, but for me this work flow is all there is, as I exclusively work for print "in house". (Till now I do not use the Web at all. That may change a tiny little bit though).
Title: SpectraView II matrices for wide gamut monitors?
Post by: probep on June 28, 2010, 08:50:23 am
Quote from: AndreaPiaggesi
Maybe I'm wrong but the MDVSENSOR2 is the custom calibrated sensor especially made for wide-gamut LCD; the MDVSENSOR is the previous version.
Excuse me, but... Have you (or anybody else) ever seen MDSVSENSOR2 with "MDSVSENSOR2" label? Once more again, i bought MDSVSENSOR2 at real NEC Store in May 2010. Did you read the FAQ from NEC?
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 28, 2010, 08:57:21 am
Quote from: probep
. Did you read the FAQ from NEC?

please point to the source. thanks in advance.
Title: SpectraView II matrices for wide gamut monitors?
Post by: probep on June 28, 2010, 09:21:27 am
Quote from: shayaweiss
please point to the source. thanks in advance.
Sorry, NEC has changed the FAQ. Older NEC colorimeter (MDSVSENSOR) was not custom calibrated and looked like http://www.superwarehouse.com/NEC_GammaCom...NSOR/ps/1552166 (http://www.superwarehouse.com/NEC_GammaComp_MD_Colorimeter_Calibrator/MDSVSENSOR/ps/1552166)
I had that one.
Title: SpectraView II matrices for wide gamut monitors?
Post by: shayaweiss on June 28, 2010, 09:27:59 am
Quote from: probep
Sorry, NEC has changed the FAQ. Older NEC colorimeter (MDSVSENSOR) was not custom calibrated and looked like http://www.superwarehouse.com/NEC_GammaCom...NSOR/ps/1552166 (http://www.superwarehouse.com/NEC_GammaComp_MD_Colorimeter_Calibrator/MDSVSENSOR/ps/1552166)
I had that one.

The one sold by B&H looks like that (and Mfr # - I think stands for Manufacturer reference # - MDSVSENSOR2 http://www.bhphotovideo.com/c/product/6136...lor_Sensor.html (http://www.bhphotovideo.com/c/product/613696-REG/NEC_MDSVSENSOR2_Custom_Calibrated_Color_Sensor.html) )! How should the new one look like? Like this: http://www.superwarehouse.com/NEC_Colorime...NSOR2/p/1563233 (http://www.superwarehouse.com/NEC_Colorimeter_Color_Calibration_Sensor/MDSVSENSOR2/p/1563233) ?


P.S.: EDIT: I called up B&H they told me it's just an old picture, the sensor is the new mdsvsensor2 straight from NEC. I ordered there, as they are the only ones to ship to my remote place.
Title: SpectraView II matrices for wide gamut monitors?
Post by: probep on June 28, 2010, 09:34:01 am
Quote from: shayaweiss
The one sold by B&H looks like that! How should the new one look like?
My new MDSVSENSOR2 (bought in May 2010) looks like http://www.flickr.com/photos/miahz/3989679014/ (http://www.flickr.com/photos/miahz/3989679014/)
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 28, 2010, 10:18:59 am
Quote from: AndreaPiaggesi
I completely agree with you but I usually spend a lot of time in RAW conversion, rating, adjustments, Photoshop editing etc ... and probably a perfect monitor with theoretical perfect colors will avoid strange color cast in photolab's prints and will help to get a correct file for your archive.
Tomorrow when you'll buy the next monitor your pictures will look almost the same, without big color shift.

Anyway I agree with you, nobody will see our perfect colors, they'll be limited to our workflow.
It hasn't been my experience 'slight' tolerance differences result in color casts.  For instance, NEC lists several pucks that are approved for use with their wide gamut monitors PROVIDED you use their SVII software.  Assuming everything, the software, the monitors, and the pucks.. are all working withing manufacturers tolerances then there will be no significant color casts.  However, if one variable is out of whack there will be color cast, and in all cases there will be small differences in the floating white point which then became unique to our closed systems and something we deal with in the rest of the color workflow such as printer profiles, paper types, the age if ink, and so on.. At least this has been my experience.

Photolabs and archives:  Having just switched to wide gamut displays with an sRGB Emulation mode.. guess what?  None of my archived images appear with the right colors in sRGB Emulation mode.  They do appear with the right colors when using the "Photo editing" or other wide gamut settings.  Of course these images never looked as they should on the web, nor did they print the way they should unless I used a printer with a Adobe98 or ProPhoto setup.. common photo labs didn't display the right colors.

Now.. thanks to the guys with black rimmed glasses and white coats at NEC I can easily switch my monitors to my current use/need.  I can go to sRGB Emulation mode and process an image through my workflow.. and it will look perfect on the web and perfect at a common sRGB lab.  Or.. I can go to Photo Editing mode and my archived images have the right color and will print on my closed system fine.  HOWEVER, if I want to use those images on the web or hav them printed at a common sRGB lab.. I need to change them.. and by this I mean more than assigning a different color profile.  I must put the monitors in sRGB Emulation mode and rework the image (some are easy, some are total reworks, depends on he depth of processing) and work the image.

ASFAIK there is no magic color space/gamut/profile that will work all around.. We have to be conscience of what we're trying to achieve and how we're trying to achieve it.. and be very careful to not jump to blame the monitor or the profile of the puck, etc, etc..  In other words we need to understand what we're doing and how to apply what we're doing to the task at hand.

To make matters more complicated.. some imaging programs and viewers.. are color managed and some are not.. and some will automatically switch the profile as you drag the image from one monitor to the other.. and some won't.  So.. we need to learn which programs are dual monitor friendly..

If someone.. say the Microsoft Team.. had color management as a priority several versions of Windows ago.. we'd probably have it right by now.  I'd guess we have several more generations of software to go until we do.  Take LR3 for instance..
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 28, 2010, 11:16:01 am
Quote from: Steve Weldon
It hasn't been my experience 'slight' tolerance differences result in color casts.

I’m not sure why some prevailing assumptions in this thread are that without the one colorimeter, the results are a color cast (whatever that really means). As I said here or in a similar post, my experiences testing the SpectraView with host software, using an EyeOne Display-2 versus an EyeOne Pro Spectrophotometer was a disconnect in about CCT 500K or so in target and measured white point. And considering that the values are kind of meaningless past a starting point, that you have to adjust to taste (to produce a visual match to the viewing booth), I don’t know why one would suspect that an instrument supported in the software would produce a “color cast”. Again, if you have no instrument, go for the mated colorimeter. If you have an instrument supported in the software, use it and season to taste. NEC didn’t support instruments that would fail to produce good final results.
Title: SpectraView II matrices for wide gamut monitors?
Post by: nilo on June 28, 2010, 12:06:03 pm
Quote from: digitaldog
NEC didn’t support instruments that would fail to produce good final results.

which device is not in the list?

Quote
*  GretagMacbeth Eye-One Display V1
    * X -Rite/GretagMacbeth Eye-One Display V2
    * X-Rite/GretagMacbeth Eye-One Monitor, Eye-One Pro.
    * X-Rite ColorMunki.
    * Monaco OptixXR (X-Rite DTP 94).
    * ColorVision/Datacolor Spyder 2.
    * Datacolor Spyder 3.

from http://www.necdisplay.com/supportcenter/mo.../compatibility/ (http://www.necdisplay.com/supportcenter/monitors/spectraview2/compatibility/)
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 28, 2010, 01:07:46 pm
Quote from: digitaldog
I’m not sure why some prevailing assumptions in this thread are that without the one colorimeter, the results are a color cast (whatever that really means). As I said here or in a similar post, my experiences testing the SpectraView with host software, using an EyeOne Display-2 versus an EyeOne Pro Spectrophotometer was a disconnect in about CCT 500K or so in target and measured white point. And considering that the values are kind of meaningless past a starting point, that you have to adjust to taste (to produce a visual match to the viewing booth), I don’t know why one would suspect that an instrument supported in the software would produce a “color cast”. Again, if you have no instrument, go for the mated colorimeter. If you have an instrument supported in the software, use it and season to taste. NEC didn’t support instruments that would fail to produce good final results.
That pretty much sums up in a single paragraph what I've been trying to say from post one..

And I suspect we can come to and live with this answer because we have enough experience with the entire color profile loop.. while the OP appears to have had a bad experience and not wanting to get burned again.. she's trying to learn everything there is to learn on the subject.. which I can't fault.  Still, she's buying an NEC wide gamut monitor and owns a 1di2 puck.. I say buy SVII online and see where it goes before worrying further.  'We' know where it will go because we have this equipment and it's working fine.. but she's reading some fairly esoteric testing and numbers which are convincing her there must be more difference than we're letting on..   There isn't.. but like I said I can't blame her for wanting to get to the bottom of it.  Its like a lens being "sharp enough.."  MTF tables are helpful, but it's your eye that determines if YOU find it sharp enough or not.  Still, some people get really hung up on every aspect of lens measurement.

Still, even from a purely financial view.. try what you have first.. She's ordering an NEC so SVII is a no brainer (or should be) and she has a 1di2 puck.. give it a try.. see if she's happy.. go from there.

I see this a lot in my workshops.. some haven't yet gained the experience to trust what they're seeing.. so they want a way to measure it and that measurement lends confidence.  We show them how to make a custom white balance setting.. and then we tell them to adjust the skin tones not to the most "correct" as measured, but the most pleasing... and bells and whistles start going off.  I suppose we were all like that as some point.
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 28, 2010, 02:25:42 pm
Quote from: LarryBaum
well spectracal claims that they actually started up production again
if they are telling the truth, then yes

They are telling the truth. I just checked with X-Rite and sure enough, they are producing the unit for some OEM’s which is good news.
Title: SpectraView II matrices for wide gamut monitors?
Post by: annamaerz on June 28, 2010, 02:55:25 pm
Quote from: digitaldog
They are telling the truth. I just check with X-Rite and sure enough, they are producing the unit for some OEM’s which is good news.

So here comes the question anew:

In the light of the information about recent restart of production of DTP94, which was not optimized for wide gamut displays, is it difficult, expensive, expensive to custom correct this puck like it was done by NEC for the i1d2? Considering its again a supported product, and again manufactured!
Title: SpectraView II matrices for wide gamut monitors?
Post by: digitaldog on June 28, 2010, 02:59:09 pm
Quote from: annamaerz
In the light of the information about recent restart of production of DTP94, which was not optimized for wide gamut displays, is it difficult, expensive, expensive to custom correct this puck? Considering its again a supported product, and again manufactured!

I don’t know. But what I do know is that originally the DTP94 was an expensive product to manufacturer and that was the primary reason it was canned when GMB and X-Rite became one (GMB had the EyeOne Display). That may be another reason why updating filters in an EyeOne Display-2 is the norm now. At some point, OEMs are going to balk at the costs. At what point does it not pay to do a custom upgrade when for perhaps similar money, you can get yourself a Spectrophotometer?
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on June 28, 2010, 04:58:37 pm
Quote from: Steve Weldon
Larry -  If you wouldn't mind terribly..  Would you give me the link on Amazon where you ordered yours?  Mine hasn't shipped yet and it would be somewhat reassuring they're sending me the right one.  There were two places that claim to have them.. The Amazon Market Place (the only place that will ship to me here overseas) and a place called Proline.

It would be somewhat irritating to pay the shipping over here, not to mention the heavy VAT and customs fees.. and get an old one..

Thanks!

Sorry but it appears they have sold out.

http://www.amazon.com/gp/product/B001P3OGZG/ref=oss_product (http://www.amazon.com/gp/product/B001P3OGZG/ref=oss_product)

When I first looked at it they said they had 3 left with more coming soon. When I ordered it a few days later they said they had 2 left. Now they are gone and they jacked the price ip and say 1-2 months (not exactly more coming soon hah).

Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on June 28, 2010, 05:03:22 pm
Quote from: probep
My new MDSVSENSOR2 (bought in May 2010) looks like http://www.flickr.com/photos/miahz/3989679014/ (http://www.flickr.com/photos/miahz/3989679014/)

There would be less confusion if they had just printed the "2" at the end. Pretty weird hah.
They did add the "custom calibrated for wide-gamut NEC displays" stamp though.
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on June 28, 2010, 05:06:22 pm
x
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on June 28, 2010, 05:08:49 pm
Quote from: ninoloss
which device is not in the list?



from http://www.necdisplay.com/supportcenter/mo.../compatibility/ (http://www.necdisplay.com/supportcenter/monitors/spectraview2/compatibility/)

Although they also say that most of those can't be used to measure the primaries and that they also have to use the factory values for max white RGB alignment as a base point so you are doing a partial calibration (although I suppose there is a chance your particular set may have the factory reference point perhaps measured better than what the NEC puck can do and then if you have a great puck perhaps you could do better, who knows, quite possibly not though)
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on June 28, 2010, 05:20:08 pm
Quote from: digitaldog
I don’t know. But what I do know is that originally the DTP94 was an expensive product to manufacturer and that was the primary reason it was canned when GMB and X-Rite became one (GMB had the EyeOne Display). That may be another reason why updating filters in an EyeOne Display-2 is the norm now. At some point, OEMs are going to balk at the costs. At what point does it not pay to do a custom upgrade when for perhaps similar money, you can get yourself a Spectrophotometer?

Supposedly NEC Europe does provide compensation for the DTP94b (as well as EIZO and Quato over there).
NEC USA said roughly along the lines of: that the had been out of production when they decided which puck to use, that the DTP94 is more expensive and they wanted to not let costs get too high (maybe they thought they were near borderline price of scaring amateurs away from the 90 and PA lines), that x-rite told them that each DTP94 revision needs it's own slightly different compensation table and that the filters are so tight in on sRGB that using a compensation table makes the instrument lose precision since they have to add such large compensation amounts externally to the puck (although then you see Quato bragging about their precision calibration system, so it's a little confusing).

Maybe the NEC puck does better than the DTP94 would do. If NEC exactingly calibrates each puck against a reference it could certainly be true on wide gamuts.
I'm curious as to whether the NEC pucks have close inter-instrument agreement and were calibrated super carefully or not (off the shelf eyeone D2 are supposed to have pretty horrid intrsument to instrument agreement). It will be interesting to how closely two NEC pucks agree (I should be able to get a chance to test that in couple weeks).
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on June 29, 2010, 01:46:01 pm
Quote from: LarryBaum
Sorry but it appears they have sold out.

http://www.amazon.com/gp/product/B001P3OGZG/ref=oss_product (http://www.amazon.com/gp/product/B001P3OGZG/ref=oss_product)

When I first looked at it they said they had 3 left with more coming soon. When I ordered it a few days later they said they had 2 left. Now they are gone and they jacked the price ip and say 1-2 months (not exactly more coming soon hah).
Thank you.  I suspect since the wide gamuts have been out for several years these pucks will all be the right thing.

It will be interesting to compare this against the three I already have..
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on July 01, 2010, 02:25:06 am
Quote from: digitaldog
I’m not sure why some prevailing assumptions in this thread are that without the one colorimeter, the results are a color cast (whatever that really means). As I said here or in a similar post, my experiences testing the SpectraView with host software, using an EyeOne Display-2 versus an EyeOne Pro Spectrophotometer was a disconnect in about CCT 500K or so in target and measured white point. And considering that the values are kind of meaningless past a starting point, that you have to adjust to taste (to produce a visual match to the viewing booth), I don’t know why one would suspect that an instrument supported in the software would produce a “color cast”. Again, if you have no instrument, go for the mated colorimeter. If you have an instrument supported in the software, use it and season to taste. NEC didn’t support instruments that would fail to produce good final results.

well 500k difference is pretty easy to spot
and it might even be more so were it to be a 500K difference plus slid up or down off-set from the standard curve
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on July 01, 2010, 12:33:33 pm
Quote from: LarryBaum
well 500k difference is pretty easy to spot
and it might even be more so were it to be a 500K difference plus slid up or down off-set from the standard curve
Could be.. depends..

An interesting note.. I've placed an order with 4 places that listed then in stock and ready to ship over the last 48 hours.. and each place canceled the order saying they were no longer in stock and they're estimated to get a new shipment in towards the 20th of July.. NEC themselves say they no longer have any in stock.

If I was one to speculate..
Title: SpectraView II matrices for wide gamut monitors?
Post by: nilo on July 01, 2010, 01:09:08 pm
Quote from: Steve Weldon
If I was one to speculate..

Maybe there wont be any until SpectraViewIII comes out with a new NEC custom filtered DTP94b  
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on July 02, 2010, 01:26:52 am
Quote from: Steve Weldon
Could be.. depends..

An interesting note.. I've placed an order with 4 places that listed then in stock and ready to ship over the last 48 hours.. and each place canceled the order saying they were no longer in stock and they're estimated to get a new shipment in towards the 20th of July.. NEC themselves say they no longer have any in stock.

If I was one to speculate..

.....I'd assume the PA-series is being phased out and replaced with the UA (Ultimate Awesome)-series and we have few week to few month old dinosaurs haha

ok, not likely

either they just had a run on people decided they needed to add the SVII kit (certainly possible) or yeah hmm maybe new puck?????????
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on July 02, 2010, 01:28:18 am
Quote from: Steve Weldon
Could be.. depends..

An interesting note.. I've placed an order with 4 places that listed then in stock and ready to ship over the last 48 hours.. and each place canceled the order saying they were no longer in stock and they're estimated to get a new shipment in towards the 20th of July.. NEC themselves say they no longer have any in stock.

If I was one to speculate..

speaking of 500K differences, I tried SVII with my DTP94b and the results seemed off. Using iColor 3.6 I roughly estimate that they are 500-600K off and it does look noticeably different (to me at least)
Title: SpectraView II matrices for wide gamut monitors?
Post by: nilo on July 02, 2010, 03:23:36 am
Quote from: LarryBaum
speaking of 500K differences, I tried SVII with my DTP94b and the results seemed off. Using iColor 3.6 I roughly estimate that they are 500-600K off and it does look noticeably different (to me at least)



How could this be? Is this the kind of user induced mistake that was mentioned here? Or is this the way a device is considered be "supported"?

BTW for me too, 500 K difference, with possibly an additional shift, is visually very off.
Title: SpectraView II matrices for wide gamut monitors?
Post by: probep on July 02, 2010, 08:18:45 am
Quote from: ninoloss
BTW for me too, 500 K difference, with possibly an additional shift, is visually very off.
Off-topic: Right, but the difference may be very different.
What do the guru think about such things (got from Argyll calibration)?:
Black level = 0.12 cd/m^2
White level = 120.27 cd/m^2
Aprox. gamma = 2.47
Contrast ratio = 1001:1
White chromaticity coordinates 0.2797, 0.3524
White    Correlated Color Temperature = 8001K, DE 2K to locus = 22.6
White Correlated Daylight Temperature = 7964K, DE 2K to locus = 21.8
White        Visual Color Temperature = 6558K, DE 2K to locus = 22.4
White     Visual Daylight Temperature = 6612K, DE 2K to locus = 21.5
8000 K vs 6560 K ?!
Title: SpectraView II matrices for wide gamut monitors?
Post by: issa on July 08, 2010, 03:19:16 pm
I have the European NEC Spectaview 2690 Reference and it comes with Spectraview Profiler a.k.a. basICColor, there was some comments on this forum that this software has some built adj matrices for wide Gamut monitors, whilst in the US this has been dealt with in a bespoke X-rite i1Display2.

I have sdone some tests using both i1match and Spectraview Profiler a.k.a. basICColor and can confirm that I more or less get the same results, including white balance and luminance.





 













Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on July 08, 2010, 07:40:18 pm
"QUESTION: Can I use the new MDSVSENSOR2 color sensor included in the new SVII-PRO-KIT with other 3rd party calibration applications?

ANSWER: Yes, however the custom calibration for NEC wide color gamut displays will not be available. Only the standard calibration is available."

It seems odd to me this is in the NEC faq and yet if I do try to use it on my other (standard gamut) displays using other software it gives pretty different results from either DTP94b probe.
Title: SpectraView II matrices for wide gamut monitors?
Post by: artobest on July 09, 2010, 08:30:00 am
I too have the European Spectraview Reference 2690, as well as a (non-NEC) i1D2. I had such problems getting neutrality with my display that I invested in a Spyder 3. That didn't work - gave me a green hue, especially in darker tones. When I contacted the BasICColor people in Germany they told me that is exactly what they would expect using that combination of colorimeter and display. Riiiight ....

After further tweaks and experimenting, I am now getting acceptable results with the i1D2. At least, the profiles validate superbly and colours are good. The important thing is my screen-to-print match is absolutely spot on.

So in a nutshell, the i1D2 works, the Spyder 3 doesn't. YMMV.
Title: SpectraView II matrices for wide gamut monitors?
Post by: nilo on July 09, 2010, 08:34:28 am
Quote from: artobest
After further tweaks and experimenting, I am now getting acceptable results with the i1D2.

Would you mind to tell us what exactly did you do?
Title: SpectraView II matrices for wide gamut monitors?
Post by: issa on July 09, 2010, 09:16:49 am
Quote from: ninoloss
Would you mind to tell us what exactly did you do?

I woukld like to understand what kind of teaks you made, also did you confirm with them whether tgheir software does wide gamut compensation? as my tests did not show any compensation v i1match software
Title: SpectraView II matrices for wide gamut monitors?
Post by: artobest on July 10, 2010, 09:24:32 am
OK, I was afraid you'd ask that!  

I was getting strong colour shifts - red and green banding - in 16-bit grayscale stepwedges in Photoshop. I was calibrating my Spectraview Reference 2690 using the supplied Spectraview-branded version of BasICColor and an HP-branded i1D2 that came with my printer.

Nothing I changed in my target settings seemed to improve matters (Generally I use 5800K, 2.2, min neutral black or 100cd/m2 and 400:1 contrast ratio), although the profiles I made validated very well in Spectraview.

In the end I set the stepwedge as my desktop to see whether the problem was at a system level or an application level. To my surprise, the wedge appeared neutral (or as near as dammit) on my desktop, so the problem seemed to lie with Photoshop. I went into my PS colour settings and changed the conversion engine from Adobe ACE to Microsoft ICM. Voila, problem solved.

The engineer at BasICColor said something about how the Microsoft ICM didn't use Black Point Compensation, which could account for the difference. I don't know whether this is a priceless gem of information or whether he was just trying to get rid of me.

Either way, the upshot is that, though I don't understand why (and I don't know why, for example, I see colour microbanding in a 16-bit grayscale gradient viewed at 600% - anybody care to explain?), I currently have passably neutral greys and very good colours, all with a non-NEC i1D2.
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on July 15, 2010, 09:00:42 am
Time to take a single bite of crow and see how well it goes down..   Not the whole crow mind you.. just the beak and claws.. yuk..

Weeelllll....   It came in today. (SVII Puck)  And I highly suspect I received a new upgraded model because 4 places I know of (and tried to order from when they listed them in stock) pulled them off the shelf at the same time.. and then mine shipped directly from NEC (via Amazon)..

Before (old puck).. lets say my target was 6506.. I'd get 6454.. or 6580..  each monitor would be on the different end therefore requiring the visual white point difference.  Never did get the 500 point differences some were talking about.

With the new puck.. I'm getting 6508 on one monitor and 6503 on the other (target 6506 (just talking about one profile, but it holds across all of them).. again, opposed.. but very close.. no visual white point adjustment needed.

I think for a single monitor in a closed system there would be absolutely no difference.  In an open system sent to others.. the differences are small enough it wouldn't be an issue either..   BUT with dual monitors side by side.. I'm really glad I bought this SVII puck.  I'm anal about them looking the same and this is a great comfort factor even if there is no real practical difference image quality wise.

Other small differences.. like the way if you have a thumbnail showing in explorer on monitor 2.. Win7 would show the thumbnail as too saturated or whatever it may be.. but if you opened it up in Windows Photo Gallery or some other color managed browser.. then it would be right.  Open a test target and zoom in so the face takes half the screen.. and slide it between monitor 1 and monitor 2.. it would be right when 100% on either monitor.. but there was a mid-point where that oversaturated/wrong look would be a factor.  Now this is no longer an issue.

So.. imo.. if you have a single monitor.. the 1di2 puck will serve you well.  If you run dual monitors the SVII puck will serve you better.
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on July 15, 2010, 06:14:19 pm
Quote from: LarryBaum
"QUESTION: Can I use the new MDSVSENSOR2 color sensor included in the new SVII-PRO-KIT with other 3rd party calibration applications?

ANSWER: Yes, however the custom calibration for NEC wide color gamut displays will not be available. Only the standard calibration is available."

It seems odd to me this is in the NEC faq and yet if I do try to use it on my other (standard gamut) displays using other software it gives pretty different results from either DTP94b probe.


Interestingly though measuring things on a NEC PA241W an i1Pro (ColorimetreHCFR) and the NEC puck (using SVII) and for sRGB primaries and white point the biggest difference between those two combos was .002 in terms of xy values and many measurements were only .000 or .001 'apart'/apart.

But NEC puck vs i1pro vs DTP94b original stock vs DTP94b new stock on an sRGB monitor using ColorimetreHCFR or CEDP had all three giving rather different results with the NEC puck being wayyyyyyyy out there. The two DTP units had the closest match.


For kicks I used CEDP + i1Pro to software calibrate the hardware sRGB mode calibration I had already done. It seemed to shift the gamma a bit off and it made the gray-scale get some green tones in the darker shades and things looked way better not building a fancy software LUT and just simply using the probe to give it a white point and perhaps primaries and letting the monitor do the rest itself.
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on July 15, 2010, 06:21:12 pm
Quote from: Steve Weldon
Time to take a single bite of crow and see how well it goes down..   Not the whole crow mind you.. just the beak and claws.. yuk..

Weeelllll....   It came in today. (SVII Puck)  And I highly suspect I received a new upgraded model because 4 places I know of (and tried to order from when they listed them in stock) pulled them off the shelf at the same time.. and then mine shipped directly from NEC (via Amazon)..

Before (old puck).. lets say my target was 6506.. I'd get 6454.. or 6580..  each monitor would be on the different end therefore requiring the visual white point difference.  Never did get the 500 point differences some were talking about.

With the new puck.. I'm getting 6508 on one monitor and 6503 on the other (target 6506 (just talking about one profile, but it holds across all of them).. again, opposed.. but very close.. no visual white point adjustment needed.

I think for a single monitor in a closed system there would be absolutely no difference.  In an open system sent to others.. the differences are small enough it wouldn't be an issue either..   BUT with dual monitors side by side.. I'm really glad I bought this SVII puck.  I'm anal about them looking the same and this is a great comfort factor even if there is no real practical difference image quality wise.

Other small differences.. like the way if you have a thumbnail showing in explorer on monitor 2.. Win7 would show the thumbnail as too saturated or whatever it may be.. but if you opened it up in Windows Photo Gallery or some other color managed browser.. then it would be right.  Open a test target and zoom in so the face takes half the screen.. and slide it between monitor 1 and monitor 2.. it would be right when 100% on either monitor.. but there was a mid-point where that oversaturated/wrong look would be a factor.  Now this is no longer an issue.

So.. imo.. if you have a single monitor.. the 1di2 puck will serve you well.  If you run dual monitors the SVII puck will serve you better.

are you talking about how evenon the same monitor just changing the target brightness or this or that a little and each time you calibrate it might give you say 6480 one time and 6565 another?
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on July 15, 2010, 07:16:58 pm
Quote from: LarryBaum
are you talking about how evenon the same monitor just changing the target brightness or this or that a little and each time you calibrate it might give you say 6480 one time and 6565 another?
I'm not sure I understand the question completely..

If you keep running the calibration on the same monitor the numbers stay close, perhaps if the target is 6500 the variance will be between 6490 and 6410.. this is on both monitors.  But I noticed something that makes sense.. I had turned on a 60 watt reading light right in front of the monitors (over my shoulder) and then the values changes 50-60 points in the negative direction.

A few things which were easier to notice with this SVII than the 1di2.. Warming it up.  A good hour and it stabilizes fully.  30 minutes gets you 50% more variance at the finer end of the measurements.  Ambient light made a big difference.  Laying out the cable without kinks, curves, and as straight as possible and away from other power cables (especially) and signal cables (to a lesser degree) helped the variances become more consistent..

It also 'sticks' to the screen.  My 1di2 didn't do that.  I still have to lean my 'device' (and old back scratcher leaning on the middle with a tv remote anchoring the other end, I imagine we all do something similar) against it to be sure it stays.. but its nice that it seems to adhere on its own.  Running other programs on the other monitor didn't make any difference UNLESS the light output was enough to affect the ambient light levels..

I played with it for a few hours trying to figure out what made it tick.. or tick faster/slower.

I do wish for the $200 it came with some sort of case to keep it clean.  I'll have to come up with something.  Caselogic makes nice inexpensive cases.. if they'd only open a Walmart here in Bangkok..   I'm a big believer in individual cases to keep things clean, coiling the cable in big loops and without kinks, etc..

I am curious if this SVII puck will work optimally with Imatch software (or better than my 1id2) on my laptops.. especially the LED backlit models with a wider gamut.. I was never happy (at all) with the 1id2/Imatch combo with laptops.. some it seemed fine, others it was obviously lacking.. hoping this will cure it.

I'm tempted to buy a Munki and some other pucks and so a more complete study of this area.  The Munki would make sense, since it makes profiles for scanners, printers, etc.. and seeing the difference between a spectrometer and a colorimeter could be interesting.. but I'm afraid there's no practical reason for the others.

I do plan on adding a P271w, possibly two, to the mix once the first or second generation kinks get worked out (providing there are any).. the wife is already complaining about the arcade look though..

I'd imagine the SVII puck will work better with the Samsung Plasma (this years model) 50" since it seems to have a wider gamut.. we'll see..   I have the Spyder tv puck/software for the dvd/blue-ray inputs.. but for the computer HDMI inputs the SVII should work better..

All kinds of testing I want to fool with  now.  I wonder if Android has an ap to profile it's screen.. ;o)
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on July 15, 2010, 07:56:31 pm
Quote from: Steve Weldon
I'm not sure I understand the question completely..

If you keep running the calibration on the same monitor the numbers stay close, perhaps if the target is 6500 the variance will be between 6490 and 6410.. this is on both monitors.  But I noticed something that makes sense.. I had turned on a 60 watt reading light right in front of the monitors (over my shoulder) and then the values changes 50-60 points in the negative direction.

A few things which were easier to notice with this SVII than the 1di2.. Warming it up.  A good hour and it stabilizes fully.  30 minutes gets you 50% more variance at the finer end of the measurements.  Ambient light made a big difference.  Laying out the cable without kinks, curves, and as straight as possible and away from other power cables (especially) and signal cables (to a lesser degree) helped the variances become more consistent..

It also 'sticks' to the screen.  My 1di2 didn't do that.  I still have to lean my 'device' (and old back scratcher leaning on the middle with a tv remote anchoring the other end, I imagine we all do something similar) against it to be sure it stays.. but its nice that it seems to adhere on its own.  Running other programs on the other monitor didn't make any difference UNLESS the light output was enough to affect the ambient light levels..

I played with it for a few hours trying to figure out what made it tick.. or tick faster/slower.

I do wish for the $200 it came with some sort of case to keep it clean.  I'll have to come up with something.  Caselogic makes nice inexpensive cases.. if they'd only open a Walmart here in Bangkok..   I'm a big believer in individual cases to keep things clean, coiling the cable in big loops and without kinks, etc..

I am curious if this SVII puck will work optimally with Imatch software (or better than my 1id2) on my laptops.. especially the LED backlit models with a wider gamut.. I was never happy (at all) with the 1id2/Imatch combo with laptops.. some it seemed fine, others it was obviously lacking.. hoping this will cure it.

I'm tempted to buy a Munki and some other pucks and so a more complete study of this area.  The Munki would make sense, since it makes profiles for scanners, printers, etc.. and seeing the difference between a spectrometer and a colorimeter could be interesting.. but I'm afraid there's no practical reason for the others.

I do plan on adding a P271w, possibly two, to the mix once the first or second generation kinks get worked out (providing there are any).. the wife is already complaining about the arcade look though..

I'd imagine the SVII puck will work better with the Samsung Plasma (this years model) 50" since it seems to have a wider gamut.. we'll see..   I have the Spyder tv puck/software for the dvd/blue-ray inputs.. but for the computer HDMI inputs the SVII should work better..

All kinds of testing I want to fool with  now.  I wonder if Android has an ap to profile it's screen.. ;o)

I don't think it will help much on the other wide gamut or specialized screens since outside of SVII it does a horrific job on the NEC PA241W (even on the NEC site they say that the meter alone does nothing on wide gamuts unless used with SVII; of course they also said it will still work fine on regular monitors outside of SVII but my results in those cases were nasty so far). I thought it might do better than a stock iD2 on sRGB screens since it has been calibrated to standard unlike an off the shelf part but using CEDP it did a horrible job on the my old sRGB monitor compared to DTP94b or i1Pro.

However, using the NEC puck with SVII on a NEC PA241W gave me numbers that were super-incredibly close to what the i1Pro gave me on that monitor which seems reassuring (both for the calibration quality they do on their re-branded pucks and the apparent ability of the i1Pro to measure at least the PA series wide gamuts pretty well despite the potential spectral spikes (the i1Pro has only 5nm resolution and sometimes there are super high and narrow spikes that it can miss and that can throw it off).

It seems to be able to calibrate the NEC screens to which it is specially matched as well as an i1Pro (and surely verifies the darker tones even better).
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on July 15, 2010, 08:04:32 pm
Quote from: LarryBaum
I don't think it will help much on the other wide gamut or specialized screens since outside of SVII it does a horrific job on the NEC PA241W (even on the NEC site they say that the meter alone does nothing on wide gamuts unless used with SVII; of course they also said it will still work fine on regular monitors outside of SVII but my results in those cases were nasty so far). I thought it might do better than a stock iD2 on sRGB screens since it has been calibrated to standard unlike an off the shelf part but using CEDP it did a horrible job on the my old sRGB monitor compared to DTP94b or i1Pro.

However, using the NEC puck with SVII on a NEC PA241W gave me numbers that were super-incredibly close to what the i1Pro gave me on that monitor which seems reassuring (both for the calibration quality they do on their re-branded pucks and the apparent ability of the i1Pro to measure at least the PA series wide gamuts pretty well despite the potential spectral spikes (the i1Pro has only 5nm resolution and sometimes there are super high and narrow spikes that it can miss and that can throw it off).

It seems to be able to calibrate the NEC screens to which it is specially matched as well as an i1Pro.
I'll probably confirm what you said. And then the next logical step is the Munki.. it should work with everything..

But yes.. I'm getting the feeling the SVII puck and SVII software and NEC SV series.. like peanut butter and chocolate..

With the dual 2690's side by side.. the way I've always set up my workstations even if they're P271w's.. the visual white point needs to match or it drives me nuts.. regardless of the colors matching (which they do).. and so far it looks like the SVII puck and software is the way to go.

I'm glad there's an easy solution.  Could be worse...

Btw.. this SVII puck came directly from NEC USA via Amazon.. via a shipping agent via USPS, a total of 7 days from the time it left NEC to in my hands in Bangkok.  Not so bad considering the agents and shippers involved.  This credit card fraud stuff is a real issue.. most places in the USA won't ship overseas making these guys necessary..
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on July 15, 2010, 10:48:06 pm
Quote from: Steve Weldon
I'll probably confirm what you said. And then the next logical step is the Munki.. it should work with everything..

But yes.. I'm getting the feeling the SVII puck and SVII software and NEC SV series.. like peanut butter and chocolate..

With the dual 2690's side by side.. the way I've always set up my workstations even if they're P271w's.. the visual white point needs to match or it drives me nuts.. regardless of the colors matching (which they do).. and so far it looks like the SVII puck and software is the way to go.

I'm glad there's an easy solution.  Could be worse...

Btw.. this SVII puck came directly from NEC USA via Amazon.. via a shipping agent via USPS, a total of 7 days from the time it left NEC to in my hands in Bangkok.  Not so bad considering the agents and shippers involved.  This credit card fraud stuff is a real issue.. most places in the USA won't ship overseas making these guys necessary..

Mine came in a box directly from amazon but it had an odd sticker sealing the box saying "Ready To Ship: Fullfilled by Amazon" but the box seemed like someone else, maybe NEC, had packed it up, it also had some sort of non-Amazon bar codes and SV kit codes attached to the box.

Anyway it was reassuring to see that they really do seem to have calibrated their pucks directly to their screens.
Title: SpectraView II matrices for wide gamut monitors?
Post by: Steve Weldon on July 15, 2010, 11:40:40 pm
Quote from: LarryBaum
Mine came in a box directly from amazon but it had an odd sticker sealing the box saying "Ready To Ship: Fullfilled by Amazon" but the box seemed like someone else, maybe NEC, had packed it up, it also had some sort of non-Amazon bar codes and SV kit codes attached to the box.

Anyway it was reassuring to see that they really do seem to have calibrated their pucks directly to their screens.
I was recently forced (well not really, but the times forced me.. ) into buying an Android phone.. and it has this nifty bar code scanner.. and you can get any number of scanning aps.. from groceries to products to books.. I should have thought about scanning the little square box plain box it came in.. there was a bar code on it.

Though, since it shipped from NEC directly perhaps they just have different labels for their drop shipping partners.

I noticed Amazon has 4-5 different places shipping this puck, from the Right Price Bros, to Amazon LTD, to NEC, all different prices.   I usually do 4-5 different searches.. I'm sure like me you've discovered the same item can have as many different prices as it does vendors.. and some stuff gets free shipping and no tax.. while the same item from another vendor gets taxed, charges shipping, and is often more expensive.

Now that I have it.. I like it.  Warmed up it is very consistent with very little variance.  For now I'm keeping it in it's bubble wrapped bag in its own drawer.. at least until I find a suitable case..    My 1id2 has been tossed into camera cases, suitcases, and dragged all over the world in a thin nylon bag an old Raptor drive came in..  Seemed pretty tough.. but I'm not taking chances with one that works this well..
Title: SpectraView II matrices for wide gamut monitors?
Post by: JeffKohn on July 16, 2010, 02:08:20 am
I have a question about the SVII software for those of you who are using it. Does it have an option to generate ICC V2 profiles that work with FireFox? That's one thing I hate about EIZO ColorNavigator is that it only generates v4 profiles.
Title: SpectraView II matrices for wide gamut monitors?
Post by: WombatHorror on July 16, 2010, 03:31:26 am
Quote from: JeffKohn
I have a question about the SVII software for those of you who are using it. Does it have an option to generate ICC V2 profiles that work with FireFox? That's one thing I hate about EIZO ColorNavigator is that it only generates v4 profiles.

The ones from SVII appear to be locked into v2 (there is no toggle that I can see) since I just switched into native gamut SV II calibration and loaded up Firefox and this webpage has all the colors looking fine as opposed to radioactive looking. Which is very nice, being locked into v4 would be a shame since Firefox would be ruined then unless you only browsed in sRGB sim mode (although it's probably best of all to browse as much as possible in that mode anyway if you truly get to be a stickler and don't want a gamma 2.2 to sRGB tone curve to have to be applied and rob a couple shades away).

Just did a quick look inside the profiles and both SVII and MultiProfiler make version 2.1 profiles (only AFAICT).
Title: SpectraView II matrices for wide gamut monitors?
Post by: probep on July 16, 2010, 03:41:08 am
Quote from: JeffKohn
I have a question about the SVII software for those of you who are using it. Does it have an option to generate ICC V2 profiles that work with FireFox? That's one thing I hate about EIZO ColorNavigator is that it only generates v4 profiles.
The SpectraView II software generates ICC v.2, matrix-based profiles only.
Title: Re: SpectraView II matrices for wide gamut monitors?
Post by: eronald on August 22, 2010, 10:06:50 pm
I think anyone who plans to color manage seriously should get a Munki or an i1 and amortize it over the next 2 displays and desktop printers she buys. The spectros should profile any display that comes on the market within the next 5 years or so, and at the very least, inter-instrument agreement should be better than for Spyder :)

Edmund

The off the shelf EyeOne-Display will work. As will other Colorimeters not filtered for this kind of unit. The target white point you ask for, and what you get will likely be off. In tests I did, the differences amounted to about CCT 500K. Its not really a huge big deal because the values you ask for are not as important as getting a White Point that produces a visual match to the print next to the display. Yes, the mated colorimeter is preferable! But its not a game changer either.
Title: Re: SpectraView II matrices for wide gamut monitors?
Post by: probep on September 10, 2010, 10:35:02 pm
BTW LaCie blue eye pro application has full support for hardware calibration (Monitor LUTs calibration) on NEC 90 series displays.
Does LaCie blue eye pro support NEC PA series displays? Anyone tested?
Title: Re: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on September 11, 2010, 01:41:25 am
No - it only works with older NEC x90 models that were rebranded by LaCie. It wasn't supporting my 3090WQXi, nor PA's.
Title: Re: SpectraView II matrices for wide gamut monitors?
Post by: probep on September 11, 2010, 01:59:28 am
No - it only works with older NEC x90 models that were rebranded by LaCie. It wasn't supporting my 3090WQXi, nor PA's.
Hm. Did you test Blue eye pro with your 3090WQXi indeed? Very strange.
I've tested it with NEC 2690WUXi2 and NEC 2490WUXi2. It does support these NECs.
Title: Re: SpectraView II matrices for wide gamut monitors?
Post by: Czornyj on September 16, 2010, 10:31:44 am
Hm. Did you test Blue eye pro with your 3090WQXi indeed? Very strange.
I've tested it with NEC 2690WUXi2 and NEC 2490WUXi2. It does support these NECs.

What Blue Eye Pro version are you referring to (I've tried 4.1.3)?
Title: Re: SpectraView II matrices for wide gamut monitors?
Post by: probep on September 16, 2010, 10:35:08 pm
What Blue Eye Pro version are you referring to (I've tried 4.1.3)?
LaCie blue eye pro v4.5.1 and LaCie blue eye pro Proof Edition v1.0.2.